Gary Yellen |
I am interested in the possibility of using a white light source (e.g. xenon
arc lamp) with a spinning disk system. We are developing fluorescent
biosensors based on various GFP derivatives, and the spectral flexibility of
this approach over multiple lasers would be very helpful. It seems as
though the Yokogawa system with the microlenses would have the best light
throughput, but the input is adapted to a singlemode fiber from a laser.
I have seen two publications describing this application (both done with
pre-CSU-X1 models):
The first (http://www.jstage.jst.go.jp/article/csf/33/1/133/_pdf) uses the
output of a mercury lamp focused into a 1-mm multimode fiber attached to the
normal light input port of the CSU. It is not clear how efficient or
inefficient this is - i.e. how much of this large fiber core is seen by the
CSU input optics. This was done with a CSU10, and I wonder if the new
beam-shaping optics on the CSU-X1 would make this better or worse.
The second (http://www.olympusconfocal.com/theory/noncoherentsources.html)
describes the use of a collimating lens and mirror, introduced into the CSU
at the location of the single excitation filter holder, to introduce light
from a liquid light guide into the optical path of the CSU, apparently after
the beam-shaping optics for the laser fiber. This approach requires a bit
more fabrication, and it is not clear to me whether this would even be
possible with the CSU-X1.
It seems (both from reading these sources and from manufacturers' specs)
that there should be plenty of light from a good xenon source. For
instance, the 300W xenon source (Lambda LS from Sutter Instruments) is
quoted as having ~500-700 mW of power in 25 nm width bands over the visible
wavelength range, and this is at the end of a 3mm liquid light guide.
I would be grateful for comments or suggestions about the feasibility of
this approach, or the best way to accomplish it.
Gary Yellen, Ph.D.
Department of Neurobiology
Harvard Medical School
|
Cameron, Lisa |
Hello -
I saw your post on the confocal listserv and thought I would let you know that our facility at Dana Farber has one of the new white light lasers (a supercontinuum laser) on the market. Ours is on the Leica SP5 confocal, so it is a scanning confocal and not the microlens based Yokogawa, but you are welcome to try it out. Leica gets the white light laser from a company called Koheras. It is delivered by a single fiber optic. I am fairly certain that they are using a fiber optic and coupler from PointSource. I'm not sure whether you can purchase a white light laser yourself directly from Koheras or any other company. Zeiss is also now marketing a white light laser that they call "In Tune", but it will most certainly be in conjunction with their laser scanning system, too. I found a listing online that advertises that you can get a white light laser from Koheras for around $20,000 at http://optics.org/cws/product/P000000270. I'm not sure of a reasonable price for just a white light laser, but Leica charges around $160,000 more if you add a white light laser to their scanning confocal. If you do end up getting one of these supercontinuum lasers from somewhere, make sure you get a service contract or some agreement for service. I have gone through 3 white light lasers in 6 months with Leica and a 4th will be installed in the next month. I'm not sure if they were just replacing the whole thing because it was easier or whether we really needed a whole new one (I have a feeling it's a combination of both). I have to say, as long as the white light laser is working, the Leica SP5 system works great. - Lisa ------------------------- Lisa Cameron, Ph.D. Director of Confocal and Light Microscopy Dana Farber Cancer Institute 44 Binney St. JF 220B Boston, MA 02115 617-582-8824 -----Original Message----- From: Confocal Microscopy List on behalf of Gary Yellen Sent: Sat 7/25/2009 10:33 AM To: [hidden email] Subject: [CONFOCALMICROSCOPY] white light source for Yokogawa SD? I am interested in the possibility of using a white light source (e.g. xenon arc lamp) with a spinning disk system. We are developing fluorescent biosensors based on various GFP derivatives, and the spectral flexibility of this approach over multiple lasers would be very helpful. It seems as though the Yokogawa system with the microlenses would have the best light throughput, but the input is adapted to a singlemode fiber from a laser. I have seen two publications describing this application (both done with pre-CSU-X1 models): The first (http://www.jstage.jst.go.jp/article/csf/33/1/133/_pdf) uses the output of a mercury lamp focused into a 1-mm multimode fiber attached to the normal light input port of the CSU. It is not clear how efficient or inefficient this is - i.e. how much of this large fiber core is seen by the CSU input optics. This was done with a CSU10, and I wonder if the new beam-shaping optics on the CSU-X1 would make this better or worse. The second (http://www.olympusconfocal.com/theory/noncoherentsources.html) describes the use of a collimating lens and mirror, introduced into the CSU at the location of the single excitation filter holder, to introduce light from a liquid light guide into the optical path of the CSU, apparently after the beam-shaping optics for the laser fiber. This approach requires a bit more fabrication, and it is not clear to me whether this would even be possible with the CSU-X1. It seems (both from reading these sources and from manufacturers' specs) that there should be plenty of light from a good xenon source. For instance, the 300W xenon source (Lambda LS from Sutter Instruments) is quoted as having ~500-700 mW of power in 25 nm width bands over the visible wavelength range, and this is at the end of a 3mm liquid light guide. I would be grateful for comments or suggestions about the feasibility of this approach, or the best way to accomplish it. Gary Yellen, Ph.D. Department of Neurobiology Harvard Medical School The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail. |
Alessandro Esposito |
In reply to this post by Gary Yellen
Dear Lisa,
I thought the SuperK by Koheras was more expensive. I do not know the american market, but I though they cost in excess of 60kUSD, which is reasonable because they could replace many lasers and it is a pulsed source useful for FLIM as well. Is the Keheras on the SP5 providing problems? I thought Keheras was guaranteeing their lasers for 2khrs imaging with a careful maintenence plan to minimize unplanned downtime periods. The truth is that supercontinuum lasers are a new technology and we'll have to wait some more years before they get an ideal commercial product. Still, if you do not rely exclusevly on them, they are very interesting sources. I am not sure that with an AOTF one would have enough power in combination with a spinning disk confocal. May be using excitation filters and using broader excitation bands would be fine. Cheers, Alessandro University of Cambridge www.quantitative-microscopy.org |
Cameron, Lisa |
I apologize - I did not mean for my previous email to go out to the whole list,
but was replying to the poster who first asked about a white light source for a Yokogawa SD, who also happens to have his lab close by my facility. I guess I forgot to delete the confocal listserv email address. Leica service has been great - both rapid in response and in getting things working. As I understand it, the problem has been due to a part failing inside the white light laser housing. Leica has said they are increasing their quality control for these parts that are not built by them. I think our situation has just been unlucky. I was only warning the previous poster that if they ended up getting a supercontinuum laser, that a service contract is highly recommended. And, yes, it is true that the laser is guaranteed for 2k hrs. But I don't think it is the actual laser itself that is failing. It is a failsafe measure that turns off the laser, as I understand it. As for the price - it is a ballpark figure of the additional cost of adding a white light laser to an SP5 AOBS system, but I want to be clear that it is only a very general price. I do not know what the current price of a laser from Koheras is or how many different models they may have. I just saw a webpage stating a price http://optics.org/cws/product/P000000270. Upon further review of this website, it looks like it is an ad for Koheras from April 4, 2006 promoting their product at the 2006 Photonics West conference, so the price may be out of date. Also, I do not know exactly what product was being advertised on this optics.org website (whether it is a similar product to what they supply to Leica). Certainly the supercontinuum technology is very exciting and should be able to replace many lasers. And let me just say, I am still happy that we bought it. As for whether a supercontinuum laser can be coupled to a Yokogawa, I agree that the power output at a single wavelength through an AOTF may not be enough for a Yokogawa. On the Leica system, we can turn on several laser lines close together to get higher power, with obviously a broader excitation. - Lisa --------------------------------------- Lisa Cameron, Ph.D. Director of Confocal and Light Microscopy Dana Farber Cancer Institute The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail. |
Free forum by Nabble | Edit this page |