Frame grabber for Hamamatsu CCD ORCA-ER

classic Classic list List threaded Threaded
♦
♦
11 messages Options
Lingqing Zhang Lingqing Zhang
Reply | Threaded
Open this post in threaded view
♦
♦
|

Frame grabber for Hamamatsu CCD ORCA-ER

Dear Lister,

Could somebody recommend a frame grabber to work with Hamamatsu digital camera ORCA-ER? Also a data acquisition software (free ones will be better) if frame grabber does not provide one.

Thank you very much for your help.

Lingqing

***********************************************
Lingqing Zhang, PhD
Nonlinear Optical Imaging Lab
Sensory Neuroscience Research Center
West Virginia University
1 Medical Center Drive
Morgantown, WV 26506
Phone: 304-293-5253
Email: [hidden email]
***********************************************
Zoltan Zoltan
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Frame grabber for Hamamatsu CCD ORCA-ER

Hi Lingqing,
 
We simply use our ORCA-ER with a firewire card via  the 1394 port; we use MicroManager for data acquisition, which is free and simple to use.
 
Zoltan 

On Fri, Jun 26, 2009 at 3:55 PM, Lingqing Zhang <[hidden email]> wrote:
Dear Lister,

Could somebody recommend a frame grabber to work with Hamamatsu digital camera ORCA-ER? Also a data acquisition software (free ones will be better) if frame grabber does not provide one.

Thank you very much for your help.

Lingqing

***********************************************
Lingqing Zhang, PhD
Nonlinear Optical Imaging Lab
Sensory Neuroscience Research Center
West Virginia University
1 Medical Center Drive
Morgantown, WV 26506
Phone: 304-293-5253
Email: [hidden email]
***********************************************



--

Zoltan Cseresnyes
Facility manager, Imaging Suite
Theresa Swayne Theresa Swayne
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Frame grabber for Hamamatsu CCD ORCA-ER

In reply to this post by Lingqing Zhang

If you don't have the Firewire connection for the Orca ER, you can use one of the Snapper (old-style PCI) or Phoenix (PCI and PCI Express) series from ActiveSilicon.co.uk. 

Hope this helps.
Theresa

On Jun 26, 2009, at 10:55 AM, Lingqing Zhang wrote:

Could somebody recommend a frame grabber to work with Hamamatsu digital camera ORCA-ER? Also a data acquisition software (free ones will be better) if frame grabber does not provide one.

------------------------------------
Theresa C. Swayne, Ph.D.
Manager, Confocal and Specialized Microscopy Shared Resource
Herbert Irving Comprehensive Cancer Center, Columbia University
1130 Saint Nicholas Ave, 222A
New York, NY 10032

212-851-4613


Haberman, Ann Haberman, Ann
Reply | Threaded
Open this post in threaded view
♦
♦
|

thoughts on dispersion compensation?

In reply to this post by Zoltan
Hello List!

I am considering upgrading my Ti:Sapphire laser to one that will
provide dispersion compensation of the beam before it hits the
microscope. I am hoping that someone who has purchased such a
pre-compensation module for their Coherent or Spectra Physics laser
might comment on their experience.

My current microscope configuration uses a fixed prism to provide
negative group velocity dispersion. If I understand correctly, this
will only tighten the pulse for a narrow range of laser wavelengths.

The pre-compensation units that are supplied by the laser
manufacturers are reputed to adjust the amount of correction
according to laser wavelength. Is this true or is the same amount of
compensation applied to all wavelengths?

Has anyone been in a position to compare a fixed prism arrangement to
one that is integrated with the laser system? Do you feel that it has
improved your ability to excite fluorophores at wavelengths that were
less efficient previously with a fixed prism, or did the improvement
seem negligible?

Thanks in advance for sharing your experience,
Ann Haberman


--
Ann Haberman, PhD
In Vivo Imaging Facility, Director

Department of Laboratory Medicine
Yale University School of Medicine
300 Cedar Street
TAC S541
New Haven, CT 06510

203-785-7349
203-785-5415 (fax)
[hidden email]
Jerry Sedgewick-2 Jerry Sedgewick-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: thoughts on dispersion compensation?

I have added a FEMTO pre-compensation unit to the laser path (no
commercial interest).  This unit moves prisms via motors to eliminate
issues with the laser moving as wavelengths are changed.  A big problem
that occurs when prisms don't move is that of the laser walking when
wavelengths are changed.

I did have a fixed set of prisms optimized at 820nm, and then purchased
a commercial unit (the FEMTO).  Certainly, it is not as convenient as
having the pre-compensation apparatus combined with the Ti:Sapphire (as
in the DeepSee) because you have to remember to move the prisms using
the control box, and one more thing to remember is easy to forget.  But,
I was able to use the existing Ti:Sapphire and provide pre-compensation.

I know that more than one company makes the motorized pre-compensation
units: I went with the advice of the faculty member who funded the
unit.  This particular unit allows one to choose the amount of
compensation at each wavelength.  I am not familiar with how Coherent or
Spectraphysics lasers provide correction, and whether or not there is
means for overriding settings, if necessary, or selecting from among a
series of choices for the level of correction.

As far as the ability to excite fluorophores, this doesn't change:  it's
the ability to achieve greater depth in the sample that changes.

For my custom multiphoton, the pre-compensation didn't make a huge
difference (about 50 - 75 extra microns in lymph nodes).  Perhaps that
is because I don't use silver mirrors; instead I use dichroics for
guiding the laser line, and so the dispersion is minimized (though most
of this occurs at lenses).  Also, one can create a spot that is smaller
than the back aperture of the objective to gain power and depth while
sacrificing resolution by adding a planoconvex lens at 500 to 1000mm
focal lengths.  Depending on what is imaged, a sacrifice of resolution
might be a reasonable trade off.

Jerry



Ann Haberman wrote:

> Hello List!
>
> I am considering upgrading my Ti:Sapphire laser to one that will
> provide dispersion compensation of the beam before it hits the
> microscope. I am hoping that someone who has purchased such a
> pre-compensation module for their Coherent or Spectra Physics laser
> might comment on their experience.
>
> My current microscope configuration uses a fixed prism to provide
> negative group velocity dispersion. If I understand correctly, this
> will only tighten the pulse for a narrow range of laser wavelengths.
>
> The pre-compensation units that are supplied by the laser
> manufacturers are reputed to adjust the amount of correction according
> to laser wavelength. Is this true or is the same amount of
> compensation applied to all wavelengths?
>
> Has anyone been in a position to compare a fixed prism arrangement to
> one that is integrated with the laser system? Do you feel that it has
> improved your ability to excite fluorophores at wavelengths that were
> less efficient previously with a fixed prism, or did the improvement
> seem negligible?
>
> Thanks in advance for sharing your experience,
> Ann Haberman
>
>


--

Jerry Sedgewick
Sedgewick Initiatives
965 Cromwell Avenue
Saint Paul, MN  55114
651-788-2261
[hidden email]
http://www.rawlight.com
http://www.jerrysedgewick.com


 



--- Get FREE High Speed Internet from USFamily.Net! -- http://www.usfamily.net/mkt-freepromo.html ---
Andreas Bruckbauer Andreas Bruckbauer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Beam pointing stability MaiTai Deepsee on Zeiss LSM 710, pre-chirp

Thanks to the combined efforts of the local Spectraphysics and Zeiss engineers the illumination profile is now centered over the whole wavelength range (700 - 1040nm) and does not wander any more as i previously reported, it is worth checkking this with a fluorescent slide.
To the pre-chirp with the spectraphysics deepsee: Yes it is possible and highly recommended to create your own table(s) for the prism position with wavelength, they have to be inside some "soft limits" though. Then everything works automatically unless the cable at the deepsee unit is unplugged (in this case you need to restart the laser).
We were able to upgrade an existing MaiTai with a deepsee unit.
The pre-chirp allows to use lower average laser power to generate the same 2-photon signal, I think it depends on the individual conditions if this really translates into more penetration depth.

best wishes

Andreas



Lingqing Zhang Lingqing Zhang
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Frame grabber for Hamamatsu CCD ORCA-ER

In reply to this post by Theresa Swayne
Theresa,
 
Sorry for so late response. Just back from Vacation.
Thank you very much for your help. I think I am going to try firewire + MicroManager combination first as Zoltan suggested.
 
Lingqing

On Fri, Jun 26, 2009 at 12:19 PM, Theresa Swayne <[hidden email]> wrote:

If you don't have the Firewire connection for the Orca ER, you can use one of the Snapper (old-style PCI) or Phoenix (PCI and PCI Express) series from ActiveSilicon.co.uk. 

Hope this helps.
Theresa

On Jun 26, 2009, at 10:55 AM, Lingqing Zhang wrote:

Could somebody recommend a frame grabber to work with Hamamatsu digital camera ORCA-ER? Also a data acquisition software (free ones will be better) if frame grabber does not provide one.

------------------------------------
Theresa C. Swayne, Ph.D.
Manager, Confocal and Specialized Microscopy Shared Resource
Herbert Irving Comprehensive Cancer Center, Columbia University
1130 Saint Nicholas Ave, 222A
New York, NY 10032

212-851-4613



Lingqing Zhang Lingqing Zhang
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Frame grabber for Hamamatsu CCD ORCA-ER

In reply to this post by Zoltan
Zoltan,,
 
Sorry for so late response. Just back from Vacation.
Great suggestion! Thank you very much for your help. 
 
Lingqing


On Fri, Jun 26, 2009 at 11:08 AM, Zoltan Cseresnyes <[hidden email]> wrote:
Hi Lingqing,
 
We simply use our ORCA-ER with a firewire card via  the 1394 port; we use MicroManager for data acquisition, which is free and simple to use.
 
Zoltan 

On Fri, Jun 26, 2009 at 3:55 PM, Lingqing Zhang <[hidden email]> wrote:
Dear Lister,

Could somebody recommend a frame grabber to work with Hamamatsu digital camera ORCA-ER? Also a data acquisition software (free ones will be better) if frame grabber does not provide one.

Thank you very much for your help.

Lingqing

***********************************************
Lingqing Zhang, PhD
Nonlinear Optical Imaging Lab
Sensory Neuroscience Research Center
West Virginia University
1 Medical Center Drive
Morgantown, WV 26506
Phone: 304-293-5253
Email: [hidden email]
***********************************************



--

Zoltan Cseresnyes
Facility manager, Imaging Suite

Nico Stuurman Nico Stuurman
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Frame grabber for Hamamatsu CCD ORCA-ER

In reply to this post by Lingqing Zhang
Hi Lingqing,

Thank you very much for your help. I think I am going to try firewire + MicroManager combination first as Zoltan suggested.

Please note that you do not really have a choice whether to use firewire or a framegrabber card.  It depends on the type of ORCA ER you have.  Hamamatsu sells cameras with firewire interface and cameras with cameralink interface (that need the Snapper or Phoenix cards).  Micro-Manager will work with both (as it uses the Hamamatsu DCAM interface that works with both).
Best,

Nico


 Lingqing

On Fri, Jun 26, 2009 at 12:19 PM, Theresa Swayne <[hidden email]> wrote:

If you don't have the Firewire connection for the Orca ER, you can use one of the Snapper (old-style PCI) or Phoenix (PCI and PCI Express) series from ActiveSilicon.co.uk. 

Hope this helps.
Theresa

On Jun 26, 2009, at 10:55 AM, Lingqing Zhang wrote:

Could somebody recommend a frame grabber to work with Hamamatsu digital camera ORCA-ER? Also a data acquisition software (free ones will be better) if frame grabber does not provide one.


Knecht, David Knecht, David
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Frame grabber for Hamamatsu CCD ORCA-ER

Nico- I have the Orca with a Snapper card and an earlier post indicated that this would not work with micro-manager.  I haven't tried it myself.  Do you know if it is true?  Dave

On Jul 6, 2009, at 2:43 PM, Nico Stuurman wrote:

Hi Lingqing,

Thank you very much for your help. I think I am going to try firewire + MicroManager combination first as Zoltan suggested.

Please note that you do not really have a choice whether to use firewire or a framegrabber card.  It depends on the type of ORCA ER you have.  Hamamatsu sells cameras with firewire interface and cameras with cameralink interface (that need the Snapper or Phoenix cards).  Micro-Manager will work with both (as it uses the Hamamatsu DCAM interface that works with both).
Best,

Nico


 Lingqing

On Fri, Jun 26, 2009 at 12:19 PM, Theresa Swayne <[hidden email]> wrote:

If you don't have the Firewire connection for the Orca ER, you can use one of the Snapper (old-style PCI) or Phoenix (PCI and PCI Express) series from ActiveSilicon.co.uk. 

Hope this helps.
Theresa

On Jun 26, 2009, at 10:55 AM, Lingqing Zhang wrote:

Could somebody recommend a frame grabber to work with Hamamatsu digital camera ORCA-ER? Also a data acquisition software (free ones will be better) if frame grabber does not provide one.



Dr. David Knecht    
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)


Nico Stuurman Nico Stuurman
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Frame grabber for Hamamatsu CCD ORCA-ER

Nico- I have the Orca with a Snapper card and an earlier post indicated that this would not work with micro-manager.  I haven't tried it myself.  Do you know if it is true?  Dave

The list of hardware supported by DCAM can be found here: http://sales.hamamatsu.com/index.php?id=13188404

It indeed looks like the Snapper card is not on that list (although not sure, best to check the exact make and model of the card you have).  Sorry for the confusion.

Best,

Nico

On Jul 6, 2009, at 2:43 PM, Nico Stuurman wrote:

Hi Lingqing,

Thank you very much for your help. I think I am going to try firewire + MicroManager combination first as Zoltan suggested.

Please note that you do not really have a choice whether to use firewire or a framegrabber card.  It depends on the type of ORCA ER you have.  Hamamatsu sells cameras with firewire interface and cameras with cameralink interface (that need the Snapper or Phoenix cards).  Micro-Manager will work with both (as it uses the Hamamatsu DCAM interface that works with both).
Best,

Nico


 Lingqing

On Fri, Jun 26, 2009 at 12:19 PM, Theresa Swayne <[hidden email]> wrote:

If you don't have the Firewire connection for the Orca ER, you can use one of the Snapper (old-style PCI) or Phoenix (PCI and PCI Express) series from ActiveSilicon.co.uk. 

Hope this helps.
Theresa

On Jun 26, 2009, at 10:55 AM, Lingqing Zhang wrote:

Could somebody recommend a frame grabber to work with Hamamatsu digital camera ORCA-ER? Also a data acquisition software (free ones will be better) if frame grabber does not provide one.



Dr. David Knecht    
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)