>Jim/Guy--
>
>It looks as if the italics didn't work--I expect
>it's a weakness of the listserver.  My
>apologies.  --Would you want to try again,
>putting your initials in front of the line?
>E.g.:
>
>XX: Firstly, shorter pulses will not make the psf
>larger.  The psf depends on the wavelength,
>nothing else (and is, in principle, infinite ...)
>The only thing that will make it appear to spread
>is if you are saturating the fluorescence at the
>centre of the psf, and this will apply
>irrespective of whether you are doing 2P or
>regular confocal.  There are all sorts of reasons
>why saturation is a bad thing which we need not
>go into here.
>
>Thanks as always for contributing,
>
>Martin
>


Thanks Martin, Never know what the server will do.

In sending it again, I have fixed some (or my) typos.

JP

Following on from the messages Jim and I posted
on the list on 'chirping' we had a lengthy
off-list exchange, and we thought that it might
be of interest to others, so here is the
discussion.  Jim is in italics, I'm in normal
type.


Guy


GC: Firstly, shorter pulses will not make the psf
larger.  The psf depends on the wavelength,
nothing else (and is, in principle, infinite ...)
The only thing that will make it appear to spread
is if you are saturating the fluorescence at the
centre of the psf, and this will apply
irrespective of whether you are doing 2P or
regular confocal.  There are all sorts of reasons
why saturation is a bad thing which we need not
go into here.

JP: I think you misunderstand me. I agree that the
"shape of the converging rays" does not depend on
the pulse length. However, these converging rays
represent a light intensity at each z plane.
Normally, we assume that this intensity only
reaches the instantaneous level needed to elicit
(much) fluorescence in the "blurry blob" near
crossover. Admittedly, this blob has fussy edges
but the does have a "z-resolution," in the form
of some Gaussian-ish response function.


GC: This has nothing to do with
multiphoton or pulse length - there is no true
zero to the confocal psf in Z so very bright
objects will intrude into the planes where they
might not (naively) be expected.   I've seen this
effect any number of times in ordinary confocal
imaging.  You will find it mentioned in my book!
The only time the psf is actually affected,
though, is if we saturate in the centre - and
this applies whatever illumination mode we are
using.  I think we really agree about this.

JP: Yes. And saturation at the peak is mentioned in my book.


However, it seems obvious to me that as the peak
power goes up, it will become high enough to
produce significant signal at some (short)
distance above or below the focus plane. I agree
that the in-focus plane will still produce more
more signal but as you mention, saturation is
possible (see duty cycle below) and I think that
the flattening caused in this way, coupled with
the effect just mentioned above, could reduce
z-res (depending on the actual peak intensity
levels).

Please recall, that my comment was addressed as a
possibility and asked if anyone had noticed this.


JP: If you shorten the pulse while keeping the power
the same you will increase all 2 and 3 photon
processes, both damaging and not.  Since 2P
follows a square law and 3P a cube law you will
change the relative proportions, as Jim says. If
you shorten the pulses and keep the peak
intensity the same you will reduce overall power
which at least will reduce heating of the sample
so there should be some benefit, without
affecting multi-photon processes.  I suspect that
what most people do is something in between.

JP: Surely, if you shorten pulses WITHOUT increasing
peak power, you will reduce average excitation
(time x intensity squared) and have less signal.
I would guess that people seldom do this (except
to the extent that, when making the adjustments
for shorter pulses, you may affect the average
power inadvertently). They try to shorten pulses
while keeping AVERAGE power the same.


GC: OK, sloppy writing on my part.  I
should have said keeping average EXCITATION the
same.  But since excitation is intensity squared,
this is NOT keeping the average power the same.
I think we are in total agreement about what
happens if they keep the average power the same.

Why is 2P more damaging?  Arguably it's not -
Vadim Dedov and I were able to measure
mitochondrial membrane potential in nerve cells
with JC1 using 2-photon excitation while
equivalent single-photon excitation killed the
cells and we couldn't measure anything.


JP: But I didn't say that "2-p is more damaging". I
said very specifically that 2-photon was more
damaging-per-excitation. But as the excitations
are confined to near the focus plane, then, on
thickish specimens, there are far fewer total
excitations and therefore less damage overall.

  And mitochondria may be an anomalous test object
as they are set up to withstand singlet O2.

GC: There are plenty of
non-mitochondrial examples.  Remember the famous
Cornell sea-urchin egg division series.  José
Feijó has shown this on many different botanical
examples.  How much damage is due to out of plane
excitation is hard to know,  though I suppose one
could use a very thin sample (film of bacteria,
for example) to test this.  But UV is so
obviously and immediately damaging to many living
cells that I find your thesis hard to sustain.


JP: Once you get past 20µm in (assuming some stain
throughout), I think that out-of-plane damage is by
far the predominant mechanism (in COnfocal), especially on
the embryoes you mention. (Jayne Sparrow and the
mouse egg that hatched?)


GC: There are lots of other examples in the
literature.  What is true is that 2P can cause
different sorts of damage.  The most extreme is
breakdown caused by the electric field, which
appears as bright flashes as you scan and
'craters' thereafter.  If you increase the peak
electric field you will naturally increase this
damage.

Another point is more subtle.  Chemical
selection rules state that in a symmetrical
molecule, 2P excitation must occur to a different
state than 1P.  This means you will not excite
the S1 state, and hence you have an enhanced rate
of inter-system crossing into a triplet state.
This is a very noticeable with fluorescein, since
it is symmetrical.  There are lots of published
spectra out there now - if a fluorochrome shows
very different 1P and 2P spectra you'd do best to
avoid it.

Finally, when we compress pulses we may not get
what we think we are getting.  Chirping gives a
pulse a strange shape, which we hope will even
out to a normal pulse after passing through our
optics.  If in fact we excite with a chirped
pulse then the peak intensity may be much high
higher than we'd calculate from the nominal pulse
length and average power.


JP: One more factor. As 2p is pulsed, the duty cycle
is usually less than 10%. This means that people
often work nearer to singlet-state saturation
when using 2photon (to get an image in the same
scan time). This means that a lot more excited
molecules are present in the very high excitation
field near the centre of the focus, and increases
the likelihood of "one-plus-one" (or maybe 2 plus
one?) overexcitation. Many smart, 2-photon folks
blame this for much of the increased
bleaching/excitation noted.

GC: Well, they may be right but I'd
like to see evidence.  In all the cases I've
looked at where people reported rapid bleaching
in 2P it's been the selection rule issue that was
obviously to blame.  People think that because
FITC is such a super-bright and relatively stable
dye it must be the label of choice.  Then they
excite at about 750nm (where it does indeed have
a peak).  And it bleaches very fast indeed.  I
pointed this out at an FOM meeting more than 10
years ago, in discussion on a paper which had
measured FITC bleach rates in 2P.  Actually if
you excite FITC at 900nm it's much more stable.

JP: I accept the points about pulse shape and sym-
metrical dyes. And Piston did note extreme differences
in bleaching rates of different dyes. (But it hindsight,
I refer to the later posting by Craig Brideau where
he found 150-200 FS worked best.)


GC: The other curious feature I've
seen in postings on this issue is that most of
these people seem to be using 'experimental'
systems with nothing (other than adjusting the
laser) to control the power reaching the sample.
So their problems are hardly surprising!  Clearly
you need to use an EOM or whatever to control the
power.   You cannot adjust a Ti-S at all
satisfactorily by adjusting the input excitation
- I know, I've tried it.

JP: Again, we agree.

--
Jim Pawley (Summer address) c/o Postmaster,
Egmont, BC, Canada, V0N-1N0 604-883-2095,
[hidden email]
--
Jim Pawley (Summer address) c/o Postmaster,
Egmont, BC, Canada, V0N-1N0 604-883-2095,
[hidden email]
Wolfgang Staroske Wolfgang Staroske
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fwd: Re: Pulse compression and in vivo imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear all,

Am 20:59, schrieb James Pawley:

> <br>JP: One more factor. As 2p is pulsed, the duty cycle
> <br>is usually less than 10%. This means that people
> <br>often work nearer to singlet-state saturation
> <br>when using 2photon (to get an image in the same
> <br>scan time). This means that a lot more excited
> <br>molecules are present in the very high excitation
> <br>field near the centre of the focus, and increases
> <br>the likelihood of "one-plus-one" (or maybe 2 plus
> <br>one?) overexcitation. Many smart, 2-photon folks
> <br>blame this for much of the increased
> <br>bleaching/excitation noted.

I would like to comment on this. In FCS Experiments we see that all
dyes, even the ones which show a strong triplet fluctuation in
one-photon excitation, show no triplet fluctuation in the case of
two-photon excitation.

Our hypothesis for that is the following. The lifetime of the triplet
state is long enough, that each molecule, which entered the triplet
state, absorbs a third IR photon, which destroys the dye molecule. So
molecules which entered the triplet state are dark from this time point on.

In imaging of course this possibility is reduced because the laser is
scanned and pixel dwell times are usually in the range or below the
triplet state lifetime (few µs), while in FCS the residence time of even
small molecules are at least 20µs.

Bye Wolfgang


--
Dr. Wolfgang Staroske

Single Molecule Specialist
Light Microscopy Facility

Technische Universität Dresden
Biotechnology Center
Tatzberg 47/49
01307 Dresden, Germany

Tel.: +49 (0) 351 463-40316
Fax.: +49 (0) 351 463-40342
E-Mail: [hidden email]
Webpage: www.biotec.tu-dresden.de
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fwd: Re: Pulse compression and in vivo imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Wolfgang,

why do you think you are causing photodestruction, as opposed to triplet
state depletion back to the ground state?

thanks,

George


On 7/14/2011 10:30 AM, Wolfgang Staroske wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear all,
>
> Am 20:59, schrieb James Pawley:
>> <br>JP: One more factor. As 2p is pulsed, the duty cycle
>> <br>is usually less than 10%. This means that people
>> <br>often work nearer to singlet-state saturation
>> <br>when using 2photon (to get an image in the same
>> <br>scan time). This means that a lot more excited
>> <br>molecules are present in the very high excitation
>> <br>field near the centre of the focus, and increases
>> <br>the likelihood of "one-plus-one" (or maybe 2 plus
>> <br>one?) overexcitation. Many smart, 2-photon folks
>> <br>blame this for much of the increased
>> <br>bleaching/excitation noted.
>
> I would like to comment on this. In FCS Experiments we see that all
> dyes, even the ones which show a strong triplet fluctuation in
> one-photon excitation, show no triplet fluctuation in the case of
> two-photon excitation.
>
> Our hypothesis for that is the following. The lifetime of the triplet
> state is long enough, that each molecule, which entered the triplet
> state, absorbs a third IR photon, which destroys the dye molecule. So
> molecules which entered the triplet state are dark from this time
> point on.
>
> In imaging of course this possibility is reduced because the laser is
> scanned and pixel dwell times are usually in the range or below the
> triplet state lifetime (few µs), while in FCS the residence time of
> even small molecules are at least 20µs.
>
> Bye Wolfgang
>
>


--


George McNamara, PhD
Analytical Imaging Core Facility
University of Miami
Wolfgang Staroske Wolfgang Staroske
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fwd: Re: Pulse compression and in vivo imaging

In reply to this post by James Pawley
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi George,

that's a good point, I haven't thought about this possibility yet.

Destruction would mean excite the molecule into higher triplet states
until its is ionized.
Triplet state depletion would be something like stimulated emission
without emission of a photon.

I would favor the first one, because this process is quantum
mechanically allowed and the energies of the IR photons should be enough
to go up the electronic states.
For the triplet-state depletion, I think the probability is as low as
for the inter-system-crossing from the singlet to the triplet state and
additionally the energy of  the triplet state is probably higher than
the energy of the IR photon. Of course that could be a
two-photon-process with enough energy than in, which would probably
happen not in the same pulse but in the next ones. On the other hand I
never heard of triplet-state depletion in one-photon-excitation and if
you get the same signal in one and two-photon excitation the
probabilities of absorption / stimulated emission (of one or two photons
respectively) should be the same.

But at all I'm not an expert in quantum mechanics, the only fact I can
state is that in the case of two-photon excitation the triplet lifetime
is either very short (below time resolution <250ns) due to triplet state
depletion or infinite due to photobleaching.

Bye Wolfgang


Am 20:59, schrieb George McNamara:

> *****
> <br>To join, leave or search the confocal microscopy listserv, go to:
> <br>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> <br>*****
> <br>
> <br>Hi Wolfgang,
> <br>
> <br>why do you think you are causing photodestruction, as opposed to
> triplet
> state depletion back to the ground state?
> <br>
> <br>thanks,
> <br>
> <br>George
> <br>
> <br>
> <br>On 7/14/2011 10:30 AM, Wolfgang Staroske wrote:
> <br><blockquote type=cite>*****
> <br>To join, leave or search the confocal microscopy listserv, go to:
> <br>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> <br>*****
> <br>
> <br>Dear all,
> <br>
> <br>Am 20:59, schrieb James Pawley:
> <br><blockquote type=cite>&lt;br&gt;JP: One more factor. As 2p is
> pulsed, the duty cycle
> <br>&lt;br&gt;is usually less than 10%. This means that people
> <br>&lt;br&gt;often work nearer to singlet-state saturation
> <br>&lt;br&gt;when using 2photon (to get an image in the same
> <br>&lt;br&gt;scan time). This means that a lot more excited
> <br>&lt;br&gt;molecules are present in the very high excitation
> <br>&lt;br&gt;field near the centre of the focus, and increases
> <br>&lt;br&gt;the likelihood of "one-plus-one" (or maybe 2 plus
> <br>&lt;br&gt;one?) overexcitation. Many smart, 2-photon folks
> <br>&lt;br&gt;blame this for much of the increased
> <br>&lt;br&gt;bleaching/excitation noted.
> <br></blockquote>
> <br>I would like to comment on this. In FCS Experiments we see that all
> dyes, even the ones which show a strong triplet fluctuation in
> one-photon excitation, show no triplet fluctuation in the case of
> two-photon excitation.
> <br>
> <br>Our hypothesis for that is the following. The lifetime of the triplet
> state is long enough, that each molecule, which entered the triplet
> state, absorbs a third IR photon, which destroys the dye molecule. So
> molecules which entered the triplet state are dark from this time
> point on.
> <br>
> <br>In imaging of course this possibility is reduced because the laser is
> scanned and pixel dwell times are usually in the range or below the
> triplet state lifetime (few µs), while in FCS the residence time of
> even small molecules are at least 20µs.
> <br>
> <br>Bye Wolfgang
> <br>
> <br>
> <br></blockquote>
> <br>
> <br></body>
> </html>
> </html>

--
Dr. Wolfgang Staroske

Single Molecule Specialist
Light Microscopy Facility

Technische Universität Dresden
Biotechnology Center
Tatzberg 47/49
01307 Dresden, Germany

Tel.: +49 (0) 351 463-40316
Fax.: +49 (0) 351 463-40342
E-Mail: [hidden email]
Webpage: www.biotec.tu-dresden.de
Guy Cox-2 Guy Cox-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fwd: Re: Pulse compression and in vivo imaging

In reply to this post by Wolfgang Staroske
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

This is really interesting.  The implication is that scanned ground-state depletion methods (which many labs are experimenting with) would not work in multiphoton.  Has anyone tried it?

                                                    Guy

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
     http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Australian Centre for Microscopy & Microanalysis,
Madsen Building F09, University of Sydney, NSW 2006

Phone +61 2 9351 3176     Fax +61 2 9351 7682
             Mobile 0413 281 861
______________________________________________
      http://www.guycox.net
 

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Wolfgang Staroske
Sent: Friday, 15 July 2011 12:30 AM
To: [hidden email]
Subject: Re: Fwd: Re: Pulse compression and in vivo imaging

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear all,

Am 20:59, schrieb James Pawley:

> <br>JP: One more factor. As 2p is pulsed, the duty cycle
> <br>is usually less than 10%. This means that people
> <br>often work nearer to singlet-state saturation
> <br>when using 2photon (to get an image in the same
> <br>scan time). This means that a lot more excited
> <br>molecules are present in the very high excitation
> <br>field near the centre of the focus, and increases
> <br>the likelihood of "one-plus-one" (or maybe 2 plus
> <br>one?) overexcitation. Many smart, 2-photon folks
> <br>blame this for much of the increased
> <br>bleaching/excitation noted.

I would like to comment on this. In FCS Experiments we see that all
dyes, even the ones which show a strong triplet fluctuation in
one-photon excitation, show no triplet fluctuation in the case of
two-photon excitation.

Our hypothesis for that is the following. The lifetime of the triplet
state is long enough, that each molecule, which entered the triplet
state, absorbs a third IR photon, which destroys the dye molecule. So
molecules which entered the triplet state are dark from this time point on.

In imaging of course this possibility is reduced because the laser is
scanned and pixel dwell times are usually in the range or below the
triplet state lifetime (few µs), while in FCS the residence time of even
small molecules are at least 20µs.

Bye Wolfgang


--
Dr. Wolfgang Staroske

Single Molecule Specialist
Light Microscopy Facility

Technische Universität Dresden
Biotechnology Center
Tatzberg 47/49
01307 Dresden, Germany

Tel.: +49 (0) 351 463-40316
Fax.: +49 (0) 351 463-40342
E-Mail: [hidden email]
Webpage: www.biotec.tu-dresden.de

-----
No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1390 / Virus Database: 1516/3763 - Release Date: 07/13/11