						James Pawley | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** >Jim/Guy-- > >It looks as if the italics didn't work--I expect >it's a weakness of the listserver. My >apologies. --Would you want to try again, >putting your initials in front of the line? >E.g.: > >XX: Firstly, shorter pulses will not make the psf >larger. The psf depends on the wavelength, >nothing else (and is, in principle, infinite ...) >The only thing that will make it appear to spread >is if you are saturating the fluorescence at the >centre of the psf, and this will apply >irrespective of whether you are doing 2P or >regular confocal. There are all sorts of reasons >why saturation is a bad thing which we need not >go into here. > >Thanks as always for contributing, > >Martin > Thanks Martin, Never know what the server will do. In sending it again, I have fixed some (or my) typos. JP Following on from the messages Jim and I posted on the list on 'chirping' we had a lengthy off-list exchange, and we thought that it might be of interest to others, so here is the discussion. Jim is in italics, I'm in normal type. Guy GC: Firstly, shorter pulses will not make the psf larger. The psf depends on the wavelength, nothing else (and is, in principle, infinite ...) The only thing that will make it appear to spread is if you are saturating the fluorescence at the centre of the psf, and this will apply irrespective of whether you are doing 2P or regular confocal. There are all sorts of reasons why saturation is a bad thing which we need not go into here. JP: I think you misunderstand me. I agree that the "shape of the converging rays" does not depend on the pulse length. However, these converging rays represent a light intensity at each z plane. Normally, we assume that this intensity only reaches the instantaneous level needed to elicit (much) fluorescence in the "blurry blob" near crossover. Admittedly, this blob has fussy edges but the does have a "z-resolution," in the form of some Gaussian-ish response function. GC: This has nothing to do with multiphoton or pulse length - there is no true zero to the confocal psf in Z so very bright objects will intrude into the planes where they might not (naively) be expected. I've seen this effect any number of times in ordinary confocal imaging. You will find it mentioned in my book! The only time the psf is actually affected, though, is if we saturate in the centre - and this applies whatever illumination mode we are using. I think we really agree about this. JP: Yes. And saturation at the peak is mentioned in my book. However, it seems obvious to me that as the peak power goes up, it will become high enough to produce significant signal at some (short) distance above or below the focus plane. I agree that the in-focus plane will still produce more more signal but as you mention, saturation is possible (see duty cycle below) and I think that the flattening caused in this way, coupled with the effect just mentioned above, could reduce z-res (depending on the actual peak intensity levels). Please recall, that my comment was addressed as a possibility and asked if anyone had noticed this. JP: If you shorten the pulse while keeping the power the same you will increase all 2 and 3 photon processes, both damaging and not. Since 2P follows a square law and 3P a cube law you will change the relative proportions, as Jim says. If you shorten the pulses and keep the peak intensity the same you will reduce overall power which at least will reduce heating of the sample so there should be some benefit, without affecting multi-photon processes. I suspect that what most people do is something in between. JP: Surely, if you shorten pulses WITHOUT increasing peak power, you will reduce average excitation (time x intensity squared) and have less signal. I would guess that people seldom do this (except to the extent that, when making the adjustments for shorter pulses, you may affect the average power inadvertently). They try to shorten pulses while keeping AVERAGE power the same. GC: OK, sloppy writing on my part. I should have said keeping average EXCITATION the same. But since excitation is intensity squared, this is NOT keeping the average power the same. I think we are in total agreement about what happens if they keep the average power the same. Why is 2P more damaging? Arguably it's not - Vadim Dedov and I were able to measure mitochondrial membrane potential in nerve cells with JC1 using 2-photon excitation while equivalent single-photon excitation killed the cells and we couldn't measure anything. JP: But I didn't say that "2-p is more damaging". I said very specifically that 2-photon was more damaging-per-excitation. But as the excitations are confined to near the focus plane, then, on thickish specimens, there are far fewer total excitations and therefore less damage overall. And mitochondria may be an anomalous test object as they are set up to withstand singlet O2. GC: There are plenty of non-mitochondrial examples. Remember the famous Cornell sea-urchin egg division series. José Feijó has shown this on many different botanical examples. How much damage is due to out of plane excitation is hard to know, though I suppose one could use a very thin sample (film of bacteria, for example) to test this. But UV is so obviously and immediately damaging to many living cells that I find your thesis hard to sustain. JP: Once you get past 20µm in (assuming some stain throughout), I think that out-of-plane damage is by far the predominant mechanism (in COnfocal), especially on the embryoes you mention. (Jayne Sparrow and the mouse egg that hatched?) GC: There are lots of other examples in the literature. What is true is that 2P can cause different sorts of damage. The most extreme is breakdown caused by the electric field, which appears as bright flashes as you scan and 'craters' thereafter. If you increase the peak electric field you will naturally increase this damage. Another point is more subtle. Chemical selection rules state that in a symmetrical molecule, 2P excitation must occur to a different state than 1P. This means you will not excite the S1 state, and hence you have an enhanced rate of inter-system crossing into a triplet state. This is a very noticeable with fluorescein, since it is symmetrical. There are lots of published spectra out there now - if a fluorochrome shows very different 1P and 2P spectra you'd do best to avoid it. Finally, when we compress pulses we may not get what we think we are getting. Chirping gives a pulse a strange shape, which we hope will even out to a normal pulse after passing through our optics. If in fact we excite with a chirped pulse then the peak intensity may be much high higher than we'd calculate from the nominal pulse length and average power. JP: One more factor. As 2p is pulsed, the duty cycle is usually less than 10%. This means that people often work nearer to singlet-state saturation when using 2photon (to get an image in the same scan time). This means that a lot more excited molecules are present in the very high excitation field near the centre of the focus, and increases the likelihood of "one-plus-one" (or maybe 2 plus one?) overexcitation. Many smart, 2-photon folks blame this for much of the increased bleaching/excitation noted. GC: Well, they may be right but I'd like to see evidence. In all the cases I've looked at where people reported rapid bleaching in 2P it's been the selection rule issue that was obviously to blame. People think that because FITC is such a super-bright and relatively stable dye it must be the label of choice. Then they excite at about 750nm (where it does indeed have a peak). And it bleaches very fast indeed. I pointed this out at an FOM meeting more than 10 years ago, in discussion on a paper which had measured FITC bleach rates in 2P. Actually if you excite FITC at 900nm it's much more stable. JP: I accept the points about pulse shape and sym- metrical dyes. And Piston did note extreme differences in bleaching rates of different dyes. (But it hindsight, I refer to the later posting by Craig Brideau where he found 150-200 FS worked best.) GC: The other curious feature I've seen in postings on this issue is that most of these people seem to be using 'experimental' systems with nothing (other than adjusting the laser) to control the power reaching the sample. So their problems are hardly surprising! Clearly you need to use an EOM or whatever to control the power. You cannot adjust a Ti-S at all satisfactorily by adjusting the input excitation - I know, I've tried it. JP: Again, we agree. -- Jim Pawley (Summer address) c/o Postmaster, Egmont, BC, Canada, V0N-1N0 604-883-2095, [hidden email] -- Jim Pawley (Summer address) c/o Postmaster, Egmont, BC, Canada, V0N-1N0 604-883-2095, [hidden email]  | 
			
 
	
	 | 
						Wolfgang Staroske | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, Am 20:59, schrieb James Pawley: > <br>JP: One more factor. As 2p is pulsed, the duty cycle > <br>is usually less than 10%. This means that people > <br>often work nearer to singlet-state saturation > <br>when using 2photon (to get an image in the same > <br>scan time). This means that a lot more excited > <br>molecules are present in the very high excitation > <br>field near the centre of the focus, and increases > <br>the likelihood of "one-plus-one" (or maybe 2 plus > <br>one?) overexcitation. Many smart, 2-photon folks > <br>blame this for much of the increased > <br>bleaching/excitation noted. I would like to comment on this. In FCS Experiments we see that all dyes, even the ones which show a strong triplet fluctuation in one-photon excitation, show no triplet fluctuation in the case of two-photon excitation. Our hypothesis for that is the following. The lifetime of the triplet state is long enough, that each molecule, which entered the triplet state, absorbs a third IR photon, which destroys the dye molecule. So molecules which entered the triplet state are dark from this time point on. In imaging of course this possibility is reduced because the laser is scanned and pixel dwell times are usually in the range or below the triplet state lifetime (few µs), while in FCS the residence time of even small molecules are at least 20µs. Bye Wolfgang -- Dr. Wolfgang Staroske Single Molecule Specialist Light Microscopy Facility Technische Universität Dresden Biotechnology Center Tatzberg 47/49 01307 Dresden, Germany Tel.: +49 (0) 351 463-40316 Fax.: +49 (0) 351 463-40342 E-Mail: [hidden email] Webpage: www.biotec.tu-dresden.de  | 
			
 
	
	 | 
						George McNamara | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Wolfgang, why do you think you are causing photodestruction, as opposed to triplet state depletion back to the ground state? thanks, George On 7/14/2011 10:30 AM, Wolfgang Staroske wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > Am 20:59, schrieb James Pawley: >> <br>JP: One more factor. As 2p is pulsed, the duty cycle >> <br>is usually less than 10%. This means that people >> <br>often work nearer to singlet-state saturation >> <br>when using 2photon (to get an image in the same >> <br>scan time). This means that a lot more excited >> <br>molecules are present in the very high excitation >> <br>field near the centre of the focus, and increases >> <br>the likelihood of "one-plus-one" (or maybe 2 plus >> <br>one?) overexcitation. Many smart, 2-photon folks >> <br>blame this for much of the increased >> <br>bleaching/excitation noted. > > I would like to comment on this. In FCS Experiments we see that all > dyes, even the ones which show a strong triplet fluctuation in > one-photon excitation, show no triplet fluctuation in the case of > two-photon excitation. > > Our hypothesis for that is the following. The lifetime of the triplet > state is long enough, that each molecule, which entered the triplet > state, absorbs a third IR photon, which destroys the dye molecule. So > molecules which entered the triplet state are dark from this time > point on. > > In imaging of course this possibility is reduced because the laser is > scanned and pixel dwell times are usually in the range or below the > triplet state lifetime (few µs), while in FCS the residence time of > even small molecules are at least 20µs. > > Bye Wolfgang > > -- George McNamara, PhD Analytical Imaging Core Facility University of Miami  | 
			
 
	
	 | 
						Wolfgang Staroske | 
| 
					
		 | 
				
					
	 
				In reply to this post by James Pawley
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi George, that's a good point, I haven't thought about this possibility yet. Destruction would mean excite the molecule into higher triplet states until its is ionized. Triplet state depletion would be something like stimulated emission without emission of a photon. I would favor the first one, because this process is quantum mechanically allowed and the energies of the IR photons should be enough to go up the electronic states. For the triplet-state depletion, I think the probability is as low as for the inter-system-crossing from the singlet to the triplet state and additionally the energy of the triplet state is probably higher than the energy of the IR photon. Of course that could be a two-photon-process with enough energy than in, which would probably happen not in the same pulse but in the next ones. On the other hand I never heard of triplet-state depletion in one-photon-excitation and if you get the same signal in one and two-photon excitation the probabilities of absorption / stimulated emission (of one or two photons respectively) should be the same. But at all I'm not an expert in quantum mechanics, the only fact I can state is that in the case of two-photon excitation the triplet lifetime is either very short (below time resolution <250ns) due to triplet state depletion or infinite due to photobleaching. Bye Wolfgang Am 20:59, schrieb George McNamara: > ***** > <br>To join, leave or search the confocal microscopy listserv, go to: > <br>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > <br>***** > <br> > <br>Hi Wolfgang, > <br> > <br>why do you think you are causing photodestruction, as opposed to > triplet > state depletion back to the ground state? > <br> > <br>thanks, > <br> > <br>George > <br> > <br> > <br>On 7/14/2011 10:30 AM, Wolfgang Staroske wrote: > <br><blockquote type=cite>***** > <br>To join, leave or search the confocal microscopy listserv, go to: > <br>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > <br>***** > <br> > <br>Dear all, > <br> > <br>Am 20:59, schrieb James Pawley: > <br><blockquote type=cite><br>JP: One more factor. As 2p is > pulsed, the duty cycle > <br><br>is usually less than 10%. This means that people > <br><br>often work nearer to singlet-state saturation > <br><br>when using 2photon (to get an image in the same > <br><br>scan time). This means that a lot more excited > <br><br>molecules are present in the very high excitation > <br><br>field near the centre of the focus, and increases > <br><br>the likelihood of "one-plus-one" (or maybe 2 plus > <br><br>one?) overexcitation. Many smart, 2-photon folks > <br><br>blame this for much of the increased > <br><br>bleaching/excitation noted. > <br></blockquote> > <br>I would like to comment on this. In FCS Experiments we see that all > dyes, even the ones which show a strong triplet fluctuation in > one-photon excitation, show no triplet fluctuation in the case of > two-photon excitation. > <br> > <br>Our hypothesis for that is the following. The lifetime of the triplet > state is long enough, that each molecule, which entered the triplet > state, absorbs a third IR photon, which destroys the dye molecule. So > molecules which entered the triplet state are dark from this time > point on. > <br> > <br>In imaging of course this possibility is reduced because the laser is > scanned and pixel dwell times are usually in the range or below the > triplet state lifetime (few µs), while in FCS the residence time of > even small molecules are at least 20µs. > <br> > <br>Bye Wolfgang > <br> > <br> > <br></blockquote> > <br> > <br></body> > </html> > </html> -- Dr. Wolfgang Staroske Single Molecule Specialist Light Microscopy Facility Technische Universität Dresden Biotechnology Center Tatzberg 47/49 01307 Dresden, Germany Tel.: +49 (0) 351 463-40316 Fax.: +49 (0) 351 463-40342 E-Mail: [hidden email] Webpage: www.biotec.tu-dresden.de  | 
			
| 
					
		 | 
				
					
	 
				In reply to this post by Wolfgang Staroske
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** This is really interesting. The implication is that scanned ground-state depletion methods (which many labs are experimenting with) would not work in multiphoton. Has anyone tried it? Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Wolfgang Staroske Sent: Friday, 15 July 2011 12:30 AM To: [hidden email] Subject: Re: Fwd: Re: Pulse compression and in vivo imaging ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, Am 20:59, schrieb James Pawley: > <br>JP: One more factor. As 2p is pulsed, the duty cycle > <br>is usually less than 10%. This means that people > <br>often work nearer to singlet-state saturation > <br>when using 2photon (to get an image in the same > <br>scan time). This means that a lot more excited > <br>molecules are present in the very high excitation > <br>field near the centre of the focus, and increases > <br>the likelihood of "one-plus-one" (or maybe 2 plus > <br>one?) overexcitation. Many smart, 2-photon folks > <br>blame this for much of the increased > <br>bleaching/excitation noted. I would like to comment on this. In FCS Experiments we see that all dyes, even the ones which show a strong triplet fluctuation in one-photon excitation, show no triplet fluctuation in the case of two-photon excitation. Our hypothesis for that is the following. The lifetime of the triplet state is long enough, that each molecule, which entered the triplet state, absorbs a third IR photon, which destroys the dye molecule. So molecules which entered the triplet state are dark from this time point on. In imaging of course this possibility is reduced because the laser is scanned and pixel dwell times are usually in the range or below the triplet state lifetime (few µs), while in FCS the residence time of even small molecules are at least 20µs. Bye Wolfgang -- Dr. Wolfgang Staroske Single Molecule Specialist Light Microscopy Facility Technische Universität Dresden Biotechnology Center Tatzberg 47/49 01307 Dresden, Germany Tel.: +49 (0) 351 463-40316 Fax.: +49 (0) 351 463-40342 E-Mail: [hidden email] Webpage: www.biotec.tu-dresden.de ----- No virus found in this message. Checked by AVG - www.avg.com Version: 10.0.1390 / Virus Database: 1516/3763 - Release Date: 07/13/11  | 
			