Indo with multiphoton question

classic Classic list List threaded Threaded
♦
♦
8 messages Options
mcammer mcammer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Indo with multiphoton question

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear multiphotoners,

I'm planning to try using Indo on the multiphoton microscope next week and was wondering whether anybody has tips on best wavelengths to use for excitation (e.g. 720 or 920 nm?) or anything else of potential use.

Thank you!

Sincerely,

Michael

________________________________________________________
Michael Cammer, Assistant Research Scientist
Skirball Institute of Biomolecular Medicine
Lab: (212) 263-3208  Cell: (914) 309-3270

</PRE>
<html>
<body>
------------------------------------------------------------<br />
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.<br />
=================================
</body>
</html>
<PRE>
Paul Herzmark Paul Herzmark
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Indo with multiphoton question

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Michael,

We do it here with the laser tuned to 720. We separate the two Indo peaks
into two different PMTs by using a 440 dichroic. The emission light going to
the short wave PMT also goes through a 400/45 bandpass filter. The light to
the long wave PMT goes through a 480/50 bandpass filter.

Hope that helps!

Paul


Paul Herzmark
Specialist
[hidden email]

Department of Molecular and Cell Biology
479 Life Science Addition
University of California, Berkeley
Berkeley, CA  94720-3200
(510) 643-9603
(510) 643-9500 fax


On Fri, Jan 7, 2011 at 9:12 AM, Cammer, Michael
<[hidden email]>wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear multiphotoners,
>
> I'm planning to try using Indo on the multiphoton microscope next week and
> was wondering whether anybody has tips on best wavelengths to use for
> excitation (e.g. 720 or 920 nm?) or anything else of potential use.
>
> Thank you!
>
> Sincerely,
>
> Michael
>
> ________________________________________________________
> Michael Cammer, Assistant Research Scientist
> Skirball Institute of Biomolecular Medicine
> Lab: (212) 263-3208  Cell: (914) 309-3270
>
> </PRE>
> <html>
> <body>
> ------------------------------------------------------------<br />
> This email message, including any attachments, is for the sole use of the
> intended recipient(s) and may contain information that is proprietary,
> confidential, and exempt from disclosure under applicable law. Any
> unauthorized review, use, disclosure, or distribution is prohibited. If you
> have received this email in error please notify the sender by return email
> and delete the original message. Please note, the recipient should check
> this email and any attachments for the presence of viruses. The organization
> accepts no liability for any damage caused by any virus transmitted by this
> email.<br />
> =================================
> </body>
> </html>
> <PRE>
>
mcammer mcammer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Indo with multiphoton question

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Thank you for the Indo responses!  They are extremely helpful I think we're set to go.
Sincerely,
Michael C.



------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================
"José A. Feijó" "José A. Feijó"
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Indo with multiphoton question

In reply to this post by Paul Herzmark
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

we did it for proof of principle, fig. 28 in
Feijó & Moreno, 2004, Imaging plant cells by two-photon
excitation,Protoplasma, 223:1-32

(also attached)

basically Paul is on spot (as usual)

Em 07-01-2011 20:17, Paul Herzmark escreveu:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Michael,
>
> We do it here with the laser tuned to 720. We separate the two Indo peaks
> into two different PMTs by using a 440 dichroic. The emission light going to
> the short wave PMT also goes through a 400/45 bandpass filter. The light to
> the long wave PMT goes through a 480/50 bandpass filter.
>
> Hope that helps!
>
> Paul
>
>
> Paul Herzmark
> Specialist
> [hidden email]
>
> Department of Molecular and Cell Biology
> 479 Life Science Addition
> University of California, Berkeley
> Berkeley, CA  94720-3200
> (510) 643-9603
> (510) 643-9500 fax
>
>
> On Fri, Jan 7, 2011 at 9:12 AM, Cammer, Michael
> <[hidden email]>wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Dear multiphotoners,
>>
>> I'm planning to try using Indo on the multiphoton microscope next week and
>> was wondering whether anybody has tips on best wavelengths to use for
>> excitation (e.g. 720 or 920 nm?) or anything else of potential use.
>>
>> Thank you!
>>
>> Sincerely,
>>
>> Michael
>>
>> ________________________________________________________
>> Michael Cammer, Assistant Research Scientist
>> Skirball Institute of Biomolecular Medicine
>> Lab: (212) 263-3208  Cell: (914) 309-3270
>>
>> </PRE>
>> <html>
>> <body>
>> ------------------------------------------------------------<br />
>> This email message, including any attachments, is for the sole use of the
>> intended recipient(s) and may contain information that is proprietary,
>> confidential, and exempt from disclosure under applicable law. Any
>> unauthorized review, use, disclosure, or distribution is prohibited. If you
>> have received this email in error please notify the sender by return email
>> and delete the original message. Please note, the recipient should check
>> this email and any attachments for the presence of viruses. The organization
>> accepts no liability for any damage caused by any virus transmitted by this
>> email.<br />
>> =================================
>> </body>
>> </html>
>> <PRE>
>>
> .
>

--


**********************************************************
Jose' A. Feijo', Prof.
----------------------------------------------------------
Dep. Biologia Vegetal, Fac.Ciencias, Universidade Lisboa
PT-1749-016 Lisboa, PORTUGAL

tel. +351.21.750.00.47/00/24, fax  +351.21.750.00.48

and/ e

Inst.Gulbenkian Ciencia, PT-2780-156 Oeiras, PORTUGAL

tel. +351.21.440.79.41/00/19, fax +351.21.440.79.70
__________________________________________________________
e.mail: [hidden email]  or  [hidden email]

URL: http://www.igc.gulbenkian.pt/research/unit/38
or   http://dbv.fc.ul.pt/users/jafeijo

**********************************************************
Mark Cannell Mark Cannell
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Indo with multiphoton question

In reply to this post by mcammer
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

We used indo-1 some time ago with some success. The details are here:

Jones, K.T., Soeller, C. & Cannell, M.B. (1998) The passage of Ca2+  
and fluorescent markers between the sperm and egg after fusion in the  
mouse. Development 125, 4627-4635.



We used 690 nm but I think a bit longer would work OK too. We also  
used 850 nm for Ca-green.



Hope this helps,



Mark

On 8/01/2011, at 6:12 AM, Cammer, Michael wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear multiphotoners,
>
> I'm planning to try using Indo on the multiphoton microscope next  
> week and was wondering whether anybody has tips on best wavelengths  
> to use for excitation (e.g. 720 or 920 nm?) or anything else of  
> potential use.
>
> Thank you!
>
> Sincerely,
>
> Michael
>
> ________________________________________________________
> Michael Cammer, Assistant Research Scientist
> Skirball Institute of Biomolecular Medicine
> Lab: (212) 263-3208  Cell: (914) 309-3270
>
> </PRE>
> <html>
> <body>
> ------------------------------------------------------------<br />
> This email message, including any attachments, is for the sole use  
> of the intended recipient(s) and may contain information that is  
> proprietary, confidential, and exempt from disclosure under  
> applicable law. Any unauthorized review, use, disclosure, or  
> distribution is prohibited. If you have received this email in error  
> please notify the sender by return email and delete the original  
> message. Please note, the recipient should check this email and any  
> attachments for the presence of viruses. The organization accepts no  
> liability for any damage caused by any virus transmitted by this  
> email.<br />
> =================================
> </body>
> </html>
> <PRE>
Alberto Diaspro Alberto Diaspro
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Indo with multiphoton question

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

we used 720
Alby

Il giorno 30/mar/2011, alle ore 21.56, Mark Cannell ha scritto:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> We used indo-1 some time ago with some success. The details are here:
>
> Jones, K.T., Soeller, C. & Cannell, M.B. (1998) The passage of Ca2+ and fluorescent markers between the sperm and egg after fusion in the mouse. Development 125, 4627-4635.
>
>
>
> We used 690 nm but I think a bit longer would work OK too. We also used 850 nm for Ca-green.
>
>
>
> Hope this helps,
>
>
>
> Mark
>
> On 8/01/2011, at 6:12 AM, Cammer, Michael wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Dear multiphotoners,
>>
>> I'm planning to try using Indo on the multiphoton microscope next week and was wondering whether anybody has tips on best wavelengths to use for excitation (e.g. 720 or 920 nm?) or anything else of potential use.
>>
>> Thank you!
>>
>> Sincerely,
>>
>> Michael
>>
>> ________________________________________________________
>> Michael Cammer, Assistant Research Scientist
>> Skirball Institute of Biomolecular Medicine
>> Lab: (212) 263-3208  Cell: (914) 309-3270
>>
>> </PRE>
>> <html>
>> <body>
>> ------------------------------------------------------------<br />
>> This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.<br />
>> =================================
>> </body>
>> </html>
>> <PRE>

 
ISTITUTO ITALIANO
DI TECNOLOGIA
 
Prof. Alberto Diaspro
Scientific Head
Nanophysics
Via Morego, 30 16163 Genova
Tel: +39-010.71.781.503
Fax +39-010-72.03.21
Mobile +39-3666719968
www.iit.it
[hidden email]

 
Roland Nitschke Roland Nitschke
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Indo with multiphoton question

In reply to this post by mcammer
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hallo Michael,

long time ago I used Indo as well as Fura-2 (right side of isobestic
point excitation) with 780 nm with quite good response to Ca.

Best regards

Roland

___________________________
Nitschke, Roland Dr.
Life Imaging Center
Centre of Systems Biology
Albert-Ludwigs-University Freiburg
Habsburgerstr.49
D-79104 Freiburg
Germany
___________________________
E-mail: [hidden email]
phone: 49-761-2032934 or 2902
fax: 49-761-2032941
http://www.imaging.uni-freiburg.de/

Am 07.01.2011 18:12, schrieb Cammer, Michael:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear multiphotoners,
>
> I'm planning to try using Indo on the multiphoton microscope next week and was wondering whether anybody has tips on best wavelengths to use for excitation (e.g. 720 or 920 nm?) or anything else of potential use.
>
> Thank you!
>
> Sincerely,
>
> Michael
>
> ________________________________________________________
> Michael Cammer, Assistant Research Scientist
> Skirball Institute of Biomolecular Medicine
> Lab: (212) 263-3208  Cell: (914) 309-3270
>
> </PRE>
> <html>
> <body>
> ------------------------------------------------------------<br />
> This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.<br />
> =================================
> </body>
> </html>
> <PRE>
>
Mark Cannell-2 Mark Cannell-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Indo with multiphoton question

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

  We went to 690nm to get a good info-1 ratio signal (Jones KT, Soeller C, Cannell MB. The passage of Ca2+ and fluorescent markers between the sperm and egg after fusion in the mouse. Development. 1998;125(23):4627–35. ) but     tuning there may be a problem. I'd say err on the short side…

 Cheers Mark


On 3/12/2012, at 3:05 PM, Roland Nitschke <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hallo Michael,
>
> long time ago I used Indo as well as Fura-2 (right side of isobestic point excitation) with 780 nm with quite good response to Ca.
>
> Best regards
>
> Roland
>
> ___________________________
> Nitschke, Roland Dr.
> Life Imaging Center
> Centre of Systems Biology
> Albert-Ludwigs-University Freiburg
> Habsburgerstr.49
> D-79104 Freiburg
> Germany
> ___________________________
> E-mail: [hidden email]
> phone: 49-761-2032934 or 2902
> fax: 49-761-2032941
> http://www.imaging.uni-freiburg.de/
>
> Am 07.01.2011 18:12, schrieb Cammer, Michael:
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Dear multiphotoners,
>>
>> I'm planning to try using Indo on the multiphoton microscope next week and was wondering whether anybody has tips on best wavelengths to use for excitation (e.g. 720 or 920 nm?) or anything else of potential use.
>>
>> Thank you!
>>
>> Sincerely,
>>
>> Michael
>>
>> ________________________________________________________
>> Michael Cammer, Assistant Research Scientist
>> Skirball Institute of Biomolecular Medicine
>> Lab: (212) 263-3208  Cell: (914) 309-3270
>>
>> </PRE>
>> <html>
>> <body>
>> ------------------------------------------------------------<br />
>> This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.<br />
>> =================================
>> </body>
>> </html>
>> <PRE>
>>

Mark  B. Cannell Ph.D. FRSNZ
Professor of Cardiac Cell Biology
School of Physiology&  Pharmacology
Medical Sciences Building
University of Bristol
Bristol
BS8 1TD UK

[hidden email]