possible crystals on a camera chip

classic Classic list List threaded Threaded
♦
♦
8 messages Options
Vita Solovyeva-2 Vita Solovyeva-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

possible crystals on a camera chip

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear All,

I would like to ask your advice about the situation we have with our Leica camera (DFC900 GT) on our our life cell Ca-Imaging system.

We noticed that we see crystals in our images (see the images for 20x and 100x, made without sample and without oil for 100x, in the attached link https://cloudstorage.elearning.uni-oldenburg.de/s/7fMgzCK3XnTGWbo ).

What I can see:

1. Crystals size does not change with magnification change (size stays the same for 20x, 40x, 63x, 100x).
2. When we turn the camera, the crystals do not rotate.
3. We can see the crystals even without objective, i.e. we remove an objective and have just ambient light shining, we still can see the crystals in the software.
4. We do not see crystals in the ocular.
5. When we opened the camera and looked at it, we could not see anything special, no matt film etc.

We asked Leica, they say that it is some crystallisation on the camera chip. The repair is about 11000€, new camera is about 16000€.

I would like to ask if you experienced a similar problem, how did you solve it?
Any ideas and suggestions are welcome!

Thank you in advance.

best wishes,
Vita


Dr. Vita Solovyeva
University of Oldenburg
Fak V- Institute of Physics
Carl von Ossietzky-Straße 9-11
26129 Oldenburg
Tel. +49-4417983547
[hidden email]
https://uol.de/fk5/elektronenmikroskopie
Emma Perry Emma Perry
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: possible crystals on a camera chip

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I know this is not quite the same but we had an issue with our Gatan camera
which is also a CMOS so it has a similar chip and the tech had us do an
annealing cycle to clean the chip, and then a new set of reference images
worked wonders to remove what noise was left. It might be worth a shot.

Good luck

Emma


On Thu, Jan 21, 2021 at 7:58 AM Vita Solovyeva <
[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Dear All,
>
> I would like to ask your advice about the situation we have with our Leica
> camera (DFC900 GT) on our our life cell Ca-Imaging system.
>
> We noticed that we see crystals in our images (see the images for 20x and
> 100x, made without sample and without oil for 100x, in the attached link
> https://cloudstorage.elearning.uni-oldenburg.de/s/7fMgzCK3XnTGWbo ).
>
> What I can see:
>
> 1. Crystals size does not change with magnification change (size stays the
> same for 20x, 40x, 63x, 100x).
> 2. When we turn the camera, the crystals do not rotate.
> 3. We can see the crystals even without objective, i.e. we remove an
> objective and have just ambient light shining, we still can see the
> crystals in the software.
> 4. We do not see crystals in the ocular.
> 5. When we opened the camera and looked at it, we could not see anything
> special, no matt film etc.
>
> We asked Leica, they say that it is some crystallisation on the camera
> chip. The repair is about 11000€, new camera is about 16000€.
>
> I would like to ask if you experienced a similar problem, how did you
> solve it?
> Any ideas and suggestions are welcome!
>
> Thank you in advance.
>
> best wishes,
> Vita
>
>
> Dr. Vita Solovyeva
> University of Oldenburg
> Fak V- Institute of Physics
> Carl von Ossietzky-Straße 9-11
> 26129 Oldenburg
> Tel. +49-4417983547
> [hidden email]
> https://uol.de/fk5/elektronenmikroskopie
>
Laurel Ballantyne Laurel Ballantyne
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: possible crystals on a camera chip

In reply to this post by Vita Solovyeva-2
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Vita,

We had a similar problem on a camera.  It turned out to be condensation which turned to ice.  We didn't get the camera repaired, but I turned off the camera cooling function and monitored the temperature. I only used it for brief periods, so this was a solution for me.

Good luck!!

Laurel

Laurel Ballantyne
Research Assistant/Lab Manager for Dr. Lindsay Fitzpatrick and Dr. Laura Wells
Department of Chemical Engineering
Queen's University
Kingston, ON
lab/office: HMRC 4-3-302
website: https://fitzpatricklab.engineering.queensu.ca/

________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Vita Solovyeva <[hidden email]>
Sent: January 21, 2021 7:48 AM
To: [hidden email] <[hidden email]>
Subject: possible crystals on a camera chip

*****
To join, leave or search the confocal microscopy listserv, go to:
https://can01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=oaeycTLnYiUgSgluezss66e252okToAV2gDqEp%2FeKO8%3D&amp;reserved=0
Post images on https://can01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=9uH0XR%2FVOKr3dcLqB1PEaVIYARJq%2Bkih9IY7sVBSWiA%3D&amp;reserved=0 and include the link in your posting.
*****

Dear All,

I would like to ask your advice about the situation we have with our Leica camera (DFC900 GT) on our our life cell Ca-Imaging system.

We noticed that we see crystals in our images (see the images for 20x and 100x, made without sample and without oil for 100x, in the attached link https://can01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fcloudstorage.elearning.uni-oldenburg.de%2Fs%2F7fMgzCK3XnTGWbo&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=ytFeP5IBj1%2FougJXBlAP6DmI7ENdJ%2F9Y4gBMx%2BuM1PA%3D&amp;reserved=0 ).

What I can see:

1. Crystals size does not change with magnification change (size stays the same for 20x, 40x, 63x, 100x).
2. When we turn the camera, the crystals do not rotate.
3. We can see the crystals even without objective, i.e. we remove an objective and have just ambient light shining, we still can see the crystals in the software.
4. We do not see crystals in the ocular.
5. When we opened the camera and looked at it, we could not see anything special, no matt film etc.

We asked Leica, they say that it is some crystallisation on the camera chip. The repair is about 11000€, new camera is about 16000€.

I would like to ask if you experienced a similar problem, how did you solve it?
Any ideas and suggestions are welcome!

Thank you in advance.

best wishes,
Vita


Dr. Vita Solovyeva
University of Oldenburg
Fak V- Institute of Physics
Carl von Ossietzky-Straße 9-11
26129 Oldenburg
Tel. +49-4417983547
[hidden email]
https://can01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fuol.de%2Ffk5%2Felektronenmikroskopie&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=gMOEnGbahvwb2oQMAX0Bg33XNNn831%2F20NZaYBSgZlI%3D&amp;reserved=0
Gerhard Holst Gerhard Holst
Reply | Threaded
Open this post in threaded view
♦
♦
|

AW: possible crystals on a camera chip

In reply to this post by Vita Solovyeva-2
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Vita,

the images look like condensation. Is the camera a cooled camera? Maybe the camera head has caught to much humdity, because if so, I mean if the scavengers are filled and/or the gas sealing was aging it might be that when switched on and the camera starts cooling the image sensor, then the humidity condensates and the images look similar.
I guess the camera needs a service to refresh the dry gas or the vacuum and dry it.

Best regards,
Gerhard





Dr. Gerhard Holst
Head of Research & Science
Business Development
+49 (0) 9441 2005 570

PCO AG, Donaupark 11, 93309 Kelheim, Germany, www.pco.de
Vorstand / Managing Board: Alexander Grünig, Luitpold Kaspar
Vorsitzender des Aufsichtsrats / Chairman of the Supervisory Board: Dr. Emil Ott
USt. ID-Nr. / VAT: DE128590843, Registergericht / Register Court: Amtsgericht Regensburg HRB 9157
Sitz der Gesellschaft / Registered Office: Kelheim



-----Ursprüngliche Nachricht-----
Von: Confocal Microscopy List <[hidden email]> Im Auftrag von Vita Solovyeva
Gesendet: Donnerstag, 21. Januar 2021 13:49
An: [hidden email]
Betreff: possible crystals on a camera chip

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear All,

I would like to ask your advice about the situation we have with our Leica camera (DFC900 GT) on our our life cell Ca-Imaging system.

We noticed that we see crystals in our images (see the images for 20x and 100x, made without sample and without oil for 100x, in the attached link https://cloudstorage.elearning.uni-oldenburg.de/s/7fMgzCK3XnTGWbo ).

What I can see:

1. Crystals size does not change with magnification change (size stays the same for 20x, 40x, 63x, 100x).
2. When we turn the camera, the crystals do not rotate.
3. We can see the crystals even without objective, i.e. we remove an objective and have just ambient light shining, we still can see the crystals in the software.
4. We do not see crystals in the ocular.
5. When we opened the camera and looked at it, we could not see anything special, no matt film etc.

We asked Leica, they say that it is some crystallisation on the camera chip. The repair is about 11000€, new camera is about 16000€.

I would like to ask if you experienced a similar problem, how did you solve it?
Any ideas and suggestions are welcome!

Thank you in advance.

best wishes,
Vita


Dr. Vita Solovyeva
University of Oldenburg
Fak V- Institute of Physics
Carl von Ossietzky-Straße 9-11
26129 Oldenburg
Tel. +49-4417983547
[hidden email]
https://uol.de/fk5/elektronenmikroskopie
Cammer, Michael-2 Cammer, Michael-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: possible crystals on a camera chip

In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

It is condensation.  https://www.flickr.com/photos/mcammer/35345268455/  We have had past discussions of this here.  Some cameras from the 1990s had to be attached to a vacuum pump periodically and resealed (I did this for someone a few years ago who was still using a Princeton Instruments camera with Windows 2000!)  But now cameras should not have the seal break.  
It needs to be fixed, you need to move the scope to a dry room, or you need a new camera.
Cheers-

Michael Cammer, Sr Research Scientist, DART Microscopy Laboratory
NYU Langone Health, 540 First Avenue, SK2 Microscopy Suite, New York, NY  10016
Office: 646-501-0567 Cell: 914-309-3270  [hidden email]  
http://nyulmc.org/micros  http://microscopynotes.com/
 



-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Laurel Ballantyne
Sent: Thursday, January 21, 2021 9:01 AM
To: [hidden email]
Subject: Re: possible crystals on a camera chip

[EXTERNAL]

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Vita,

We had a similar problem on a camera.  It turned out to be condensation which turned to ice.  We didn't get the camera repaired, but I turned off the camera cooling function and monitored the temperature. I only used it for brief periods, so this was a solution for me.

Good luck!!

Laurel

Laurel Ballantyne
Research Assistant/Lab Manager for Dr. Lindsay Fitzpatrick and Dr. Laura Wells Department of Chemical Engineering Queen's University Kingston, ON
lab/office: HMRC 4-3-302
website: https://fitzpatricklab.engineering.queensu.ca/

________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Vita Solovyeva <[hidden email]>
Sent: January 21, 2021 7:48 AM
To: [hidden email] <[hidden email]>
Subject: possible crystals on a camera chip

*****
To join, leave or search the confocal microscopy listserv, go to:
https://can01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=oaeycTLnYiUgSgluezss66e252okToAV2gDqEp%2FeKO8%3D&amp;reserved=0
Post images on https://can01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=9uH0XR%2FVOKr3dcLqB1PEaVIYARJq%2Bkih9IY7sVBSWiA%3D&amp;reserved=0 and include the link in your posting.
*****

Dear All,

I would like to ask your advice about the situation we have with our Leica camera (DFC900 GT) on our our life cell Ca-Imaging system.

We noticed that we see crystals in our images (see the images for 20x and 100x, made without sample and without oil for 100x, in the attached link https://can01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fcloudstorage.elearning.uni-oldenburg.de%2Fs%2F7fMgzCK3XnTGWbo&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=ytFeP5IBj1%2FougJXBlAP6DmI7ENdJ%2F9Y4gBMx%2BuM1PA%3D&amp;reserved=0 ).

What I can see:

1. Crystals size does not change with magnification change (size stays the same for 20x, 40x, 63x, 100x).
2. When we turn the camera, the crystals do not rotate.
3. We can see the crystals even without objective, i.e. we remove an objective and have just ambient light shining, we still can see the crystals in the software.
4. We do not see crystals in the ocular.
5. When we opened the camera and looked at it, we could not see anything special, no matt film etc.

We asked Leica, they say that it is some crystallisation on the camera chip. The repair is about 11000€, new camera is about 16000€.

I would like to ask if you experienced a similar problem, how did you solve it?
Any ideas and suggestions are welcome!

Thank you in advance.

best wishes,
Vita


Dr. Vita Solovyeva
University of Oldenburg
Fak V- Institute of Physics
Carl von Ossietzky-Straße 9-11
26129 Oldenburg
Tel. +49-4417983547
[hidden email]
https://can01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fuol.de%2Ffk5%2Felektronenmikroskopie&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=gMOEnGbahvwb2oQMAX0Bg33XNNn831%2F20NZaYBSgZlI%3D&amp;reserved=0

------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================

Colin Haig Colin Haig
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: possible crystals on a camera chip

In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Vita,
The "crystals" are very sharply focused, so it is not on the camera sensor itself.  It is somewhere further down the optical path.
Ice and frost usually have more definitely fractal edges in my experience.  Perhaps this is forming around fine particles somewhere in the optical setup.
A dessicant to dry it out and heating to room temperature or a bit above might help.
Check if your camera has a rechargeable desisccant cartridge or replaceable desiccant, in case it is forming on an optical window several mm in front of the sensor.
Cleaning may be needed as well.
Good luck,
Colin

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Laurel Ballantyne
Sent: January 21, 2021 9:01 AM
To: [hidden email]
Subject: Re: possible crystals on a camera chip

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Vita,

We had a similar problem on a camera.  It turned out to be condensation which turned to ice.  We didn't get the camera repaired, but I turned off the camera cooling function and monitored the temperature. I only used it for brief periods, so this was a solution for me.

Good luck!!

Laurel

Laurel Ballantyne
Research Assistant/Lab Manager for Dr. Lindsay Fitzpatrick and Dr. Laura Wells Department of Chemical Engineering Queen's University Kingston, ON
lab/office: HMRC 4-3-302
website: https://fitzpatricklab.engineering.queensu.ca/

________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Vita Solovyeva <[hidden email]>
Sent: January 21, 2021 7:48 AM
To: [hidden email] <[hidden email]>
Subject: possible crystals on a camera chip

*****
To join, leave or search the confocal microscopy listserv, go to:
https://can01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=oaeycTLnYiUgSgluezss66e252okToAV2gDqEp%2FeKO8%3D&amp;reserved=0
Post images on https://can01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=9uH0XR%2FVOKr3dcLqB1PEaVIYARJq%2Bkih9IY7sVBSWiA%3D&amp;reserved=0 and include the link in your posting.
*****

Dear All,

I would like to ask your advice about the situation we have with our Leica camera (DFC900 GT) on our our life cell Ca-Imaging system.

We noticed that we see crystals in our images (see the images for 20x and 100x, made without sample and without oil for 100x, in the attached link https://can01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fcloudstorage.elearning.uni-oldenburg.de%2Fs%2F7fMgzCK3XnTGWbo&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=ytFeP5IBj1%2FougJXBlAP6DmI7ENdJ%2F9Y4gBMx%2BuM1PA%3D&amp;reserved=0 ).

What I can see:

1. Crystals size does not change with magnification change (size stays the same for 20x, 40x, 63x, 100x).
2. When we turn the camera, the crystals do not rotate.
3. We can see the crystals even without objective, i.e. we remove an objective and have just ambient light shining, we still can see the crystals in the software.
4. We do not see crystals in the ocular.
5. When we opened the camera and looked at it, we could not see anything special, no matt film etc.

We asked Leica, they say that it is some crystallisation on the camera chip. The repair is about 11000€, new camera is about 16000€.

I would like to ask if you experienced a similar problem, how did you solve it?
Any ideas and suggestions are welcome!

Thank you in advance.

best wishes,
Vita


Dr. Vita Solovyeva
University of Oldenburg
Fak V- Institute of Physics
Carl von Ossietzky-Straße 9-11
26129 Oldenburg
Tel. +49-4417983547
[hidden email]
https://can01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fuol.de%2Ffk5%2Felektronenmikroskopie&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=gMOEnGbahvwb2oQMAX0Bg33XNNn831%2F20NZaYBSgZlI%3D&amp;reserved=0
G. Esteban Fernandez G. Esteban Fernandez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: possible crystals on a camera chip

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

That’s what it looks like when the condensate is on the sensor.

For what it's worth, we had this issue with a 6 year old Hamamatsu Orca
Flash4.0 LT (albeit with a smaller area of condensation in the center of
the chip) and they replaced it for $3K. Seems reasonable to me; equivalent
to $500/yr extended warranty I’ve been quoted on similar cameras from other
vendors.

-Esteban

On Thu, Jan 21, 2021 at 7:22 AM Colin Haig <[hidden email]>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Vita,
> The "crystals" are very sharply focused, so it is not on the camera sensor
> itself.  It is somewhere further down the optical path.
> Ice and frost usually have more definitely fractal edges in my
> experience.  Perhaps this is forming around fine particles somewhere in the
> optical setup.
> A dessicant to dry it out and heating to room temperature or a bit above
> might help.
> Check if your camera has a rechargeable desisccant cartridge or
> replaceable desiccant, in case it is forming on an optical window several
> mm in front of the sensor.
> Cleaning may be needed as well.
> Good luck,
> Colin
>
> -----Original Message-----
> From: Confocal Microscopy List <[hidden email]> On
> Behalf Of Laurel Ballantyne
> Sent: January 21, 2021 9:01 AM
> To: [hidden email]
> Subject: Re: possible crystals on a camera chip
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Vita,
>
> We had a similar problem on a camera.  It turned out to be condensation
> which turned to ice.  We didn't get the camera repaired, but I turned off
> the camera cooling function and monitored the temperature. I only used it
> for brief periods, so this was a solution for me.
>
> Good luck!!
>
> Laurel
>
> Laurel Ballantyne
> Research Assistant/Lab Manager for Dr. Lindsay Fitzpatrick and Dr. Laura
> Wells Department of Chemical Engineering Queen's University Kingston, ON
> lab/office: HMRC 4-3-302
> website: https://fitzpatricklab.engineering.queensu.ca/
>
> ________________________________
> From: Confocal Microscopy List <[hidden email]> on
> behalf of Vita Solovyeva <[hidden email]>
> Sent: January 21, 2021 7:48 AM
> To: [hidden email] <[hidden email]>
> Subject: possible crystals on a camera chip
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
> https://can01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=oaeycTLnYiUgSgluezss66e252okToAV2gDqEp%2FeKO8%3D&amp;reserved=0
> Post images on
> https://can01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=9uH0XR%2FVOKr3dcLqB1PEaVIYARJq%2Bkih9IY7sVBSWiA%3D&amp;reserved=0
> and include the link in your posting.
> *****
>
> Dear All,
>
> I would like to ask your advice about the situation we have with our Leica
> camera (DFC900 GT) on our our life cell Ca-Imaging system.
>
> We noticed that we see crystals in our images (see the images for 20x and
> 100x, made without sample and without oil for 100x, in the attached link
> https://can01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fcloudstorage.elearning.uni-oldenburg.de%2Fs%2F7fMgzCK3XnTGWbo&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=ytFeP5IBj1%2FougJXBlAP6DmI7ENdJ%2F9Y4gBMx%2BuM1PA%3D&amp;reserved=0
> ).
>
> What I can see:
>
> 1. Crystals size does not change with magnification change (size stays the
> same for 20x, 40x, 63x, 100x).
> 2. When we turn the camera, the crystals do not rotate.
> 3. We can see the crystals even without objective, i.e. we remove an
> objective and have just ambient light shining, we still can see the
> crystals in the software.
> 4. We do not see crystals in the ocular.
> 5. When we opened the camera and looked at it, we could not see anything
> special, no matt film etc.
>
> We asked Leica, they say that it is some crystallisation on the camera
> chip. The repair is about 11000€, new camera is about 16000€.
>
> I would like to ask if you experienced a similar problem, how did you
> solve it?
> Any ideas and suggestions are welcome!
>
> Thank you in advance.
>
> best wishes,
> Vita
>
>
> Dr. Vita Solovyeva
> University of Oldenburg
> Fak V- Institute of Physics
> Carl von Ossietzky-Straße 9-11
> 26129 Oldenburg
> Tel. +49-4417983547
> [hidden email]
>
> https://can01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fuol.de%2Ffk5%2Felektronenmikroskopie&amp;data=04%7C01%7Cballantl%40QUEENSU.CA%7C5ed7648d33a646d5bffe08d8be0c5457%7Cd61ecb3b38b142d582c4efb2838b925c%7C1%7C0%7C637468307463400452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C2000&amp;sdata=gMOEnGbahvwb2oQMAX0Bg33XNNn831%2F20NZaYBSgZlI%3D&amp;reserved=0
>
Praju Vikas Anekal Praju Vikas Anekal
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: possible crystals on a camera chip

In reply to this post by Vita Solovyeva-2
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Vita,

We had a similar problem with our CMOS on our High content system.
I concur with a few other replies that it is likely to be condensation.

In our case, the camera seal had eroded and it let moisture into the sensor array area.
The engineer said that resealing etc was not an option and a new camera was needed($$$)
We disabled the environmental control and left the camera on 24/7 to let the fan drive out the moisture.
Fortunately, this approach has worked so far for us.

Cheers

Yours sincerely,

Praju Vikas Anekal. Ph.D.
Biomed Imaging Microscopist/BioImage Analyst, Biomedical Imaging Research Unit.
Faculty of Medical and Health Sciences, The University of Auckland.
E-Mail : [hidden email]<mailto:[hidden email]> , Ext : 87831

From: Confocal Microscopy List <[hidden email]> On Behalf Of Vita Solovyeva
Sent: Friday, 22 January 2021 1:49 AM
To: [hidden email]
Subject: possible crystals on a camera chip

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>
Post images on http://www.imgur.com<http://www.imgur.com> and include the link in your posting.
*****

Dear All,

I would like to ask your advice about the situation we have with our Leica camera (DFC900 GT) on our our life cell Ca-Imaging system.

We noticed that we see crystals in our images (see the images for 20x and 100x, made without sample and without oil for 100x, in the attached link https://cloudstorage.elearning.uni-oldenburg.de/s/7fMgzCK3XnTGWbo<https://cloudstorage.elearning.uni-oldenburg.de/s/7fMgzCK3XnTGWbo> ).

What I can see:

1. Crystals size does not change with magnification change (size stays the same for 20x, 40x, 63x, 100x).
2. When we turn the camera, the crystals do not rotate.
3. We can see the crystals even without objective, i.e. we remove an objective and have just ambient light shining, we still can see the crystals in the software.
4. We do not see crystals in the ocular.
5. When we opened the camera and looked at it, we could not see anything special, no matt film etc.

We asked Leica, they say that it is some crystallisation on the camera chip. The repair is about 11000€, new camera is about 16000€.

I would like to ask if you experienced a similar problem, how did you solve it?
Any ideas and suggestions are welcome!

Thank you in advance.

best wishes,
Vita