Bernhard Kuhn |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear All, We have an Olympus IX81 with a motorized stage Prior Proscan II, microscopy control unit IX2-USB and Olympus Laser- Based Autofocus System (IX2-ZDC). The camera is a Hamamatsu EM-CCD C9100-02, controlled with A3472-07. We are using this unit for multi-location, multi-time microscopy ~30 locations for ~150 acquisitions per experiment. What software and operating system do people recommend to control this setup? Is there anyone with a similar setup who can make software recommendations? Thank you and Happy New Year, Bernhard Kuhn |
|
John Oreopoulos |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Check out MicroManager when you get a chance. John Oreopoulos On 2011-12-28, at 2:19 PM, Bernhard Kuhn <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear All, > > We have an Olympus IX81 with a motorized stage Prior Proscan II, microscopy > control unit IX2-USB and Olympus Laser- Based Autofocus System (IX2-ZDC). > > The camera is a Hamamatsu EM-CCD C9100-02, controlled with A3472-07. > > We are using this unit for multi-location, multi-time microscopy ~30 > locations for ~150 acquisitions per experiment. > > What software and operating system do people recommend to control this > setup? > > Is there anyone with a similar setup who can make software recommendations? > > Thank you and Happy New Year, > > Bernhard Kuhn |
|
Gary G. Li-2 |
|
|
In reply to this post by Bernhard Kuhn
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Bernhard Kuhn, Discuss with a technical support of Molecular Devices to see if their MetaMorph fits your requirements. http://www.moleculardevices.com/ Happy Holidays! Gary G Li, PhD Microscopy Facility Manager NIDA/NIH Baltimore, MD 21224 On Wed, Dec 28, 2011 at 2:19 PM, Bernhard Kuhn <[hidden email] > wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear All, > > We have an Olympus IX81 with a motorized stage Prior Proscan II, microscopy > control unit IX2-USB and Olympus Laser- Based Autofocus System (IX2-ZDC). > > The camera is a Hamamatsu EM-CCD C9100-02, controlled with A3472-07. > > We are using this unit for multi-location, multi-time microscopy ~30 > locations for ~150 acquisitions per experiment. > > What software and operating system do people recommend to control this > setup? > > Is there anyone with a similar setup who can make software recommendations? > > Thank you and Happy New Year, > > Bernhard Kuhn > -- Gary G. Li, Ph.D. |
|
|
In reply to this post by John Oreopoulos
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** > Check out MicroManager when you get a chance. > I concur. that is where things are heading, and your only bet if you want to squeeze out everything your system can do I always recommend linux as otherwise everything seems to be about handling your virus infected windows computer. unfortunately I don't think your hardware is supported on this OS yet (please put some pressure on olympus) /Johan > > John Oreopoulos > > On 2011-12-28, at 2:19 PM, Bernhard Kuhn <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Dear All, > > > > We have an Olympus IX81 with a motorized stage Prior Proscan II, > microscopy > > control unit IX2-USB and Olympus Laser- Based Autofocus System (IX2-ZDC). > > > > The camera is a Hamamatsu EM-CCD C9100-02, controlled with A3472-07. > > > > We are using this unit for multi-location, multi-time microscopy ~30 > > locations for ~150 acquisitions per experiment. > > > > What software and operating system do people recommend to control this > > setup? > > > > Is there anyone with a similar setup who can make software > recommendations? > > > > Thank you and Happy New Year, > > > > Bernhard Kuhn > -- ----------------------------------------------------------- Johan Henriksson PhD student, Karolinska Institutet http://mahogny.areta.org http://www.endrov.net |
|
Edelmann, Richard E. Dr. |
|
|
In reply to this post by Bernhard Kuhn
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Bernhard: A couple of other options to follow up on. Image Pro Plus (www.mediacy.com) does directly control the IX-81, The Prior Stages, and has very flexible macro programing controls fro allowing the slide scanning. I do not know about the Hamamatsu interface. Secondly, Slide-Book ( www.intelligent-imaging.com) will also control the IX-81. We have an older copy running deconvolution on one of our IX-81's. Richard E. Edelmann, Ph.D. Center for Advanced Microscopy & Imaging, Director Miami University Oxford, OH 45056 Ph: 513.529.5712 www.cami.muohio.edu ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Bernhard Kuhn [[hidden email]] Sent: Wednesday, December 28, 2011 2:19 PM To: [hidden email] Subject: software to control Olympus IX81 for video microscopy ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear All, We have an Olympus IX81 with a motorized stage Prior Proscan II, microscopy control unit IX2-USB and Olympus Laser- Based Autofocus System (IX2-ZDC). The camera is a Hamamatsu EM-CCD C9100-02, controlled with A3472-07. We are using this unit for multi-location, multi-time microscopy ~30 locations for ~150 acquisitions per experiment. What software and operating system do people recommend to control this setup? Is there anyone with a similar setup who can make software recommendations? Thank you and Happy New Year, Bernhard Kuhn |
|
Paul Rigby-2 |
|
|
In reply to this post by John Oreopoulos
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All, I have also been trying to control an IX81 with almost the identical hardware to Bernhard (Prior stage, ZDC etc) using Micromanager. However, I have been unable to get our computer (and Micromanager) to successfully talk with the Olympus control box via the serial port. I suspect the serial port settings. If anyone has done this successfully, can you please let me/us know what settings you used or what has worked best. Many thanks. Paul Assoc. Prof. Paul Rigby Centre for Microscopy, Characterisation & Analysis (M510) The University of Western Australia 35 Stirling Highway Crawley WA 6007 Australia -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of John Oreopoulos Sent: Thursday, 29 December 2011 4:37 AM To: [hidden email] Subject: Re: software to control Olympus IX81 for video microscopy ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Check out MicroManager when you get a chance. John Oreopoulos On 2011-12-28, at 2:19 PM, Bernhard Kuhn <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear All, > > We have an Olympus IX81 with a motorized stage Prior Proscan II, microscopy > control unit IX2-USB and Olympus Laser- Based Autofocus System (IX2-ZDC). > > The camera is a Hamamatsu EM-CCD C9100-02, controlled with A3472-07. > > We are using this unit for multi-location, multi-time microscopy ~30 > locations for ~150 acquisitions per experiment. > > What software and operating system do people recommend to control this > setup? > > Is there anyone with a similar setup who can make software recommendations? > > Thank you and Happy New Year, > > Bernhard Kuhn |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I had trouble with the IX81 once and solved it by increasing the timeout (in the hardware config). does this help? my timeout setting was probably higher than listed here: http://valelab.ucsf.edu/~MM/MMwiki/index.php/OlympusIX81/BX61 /Johan On Thu, Jan 5, 2012 at 4:55 AM, Paul Rigby <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi All, > I have also been trying to control an IX81 with almost the identical > hardware to Bernhard (Prior stage, ZDC etc) using Micromanager. However, I > have been unable to get our computer (and Micromanager) to successfully > talk with the Olympus control box via the serial port. I suspect the serial > port settings. If anyone has done this successfully, can you please let > me/us know what settings you used or what has worked best. > > Many thanks. > Paul > > Assoc. Prof. Paul Rigby > Centre for Microscopy, Characterisation & Analysis (M510) > The University of Western Australia > 35 Stirling Highway > Crawley WA 6007 > Australia > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of John Oreopoulos > Sent: Thursday, 29 December 2011 4:37 AM > To: [hidden email] > Subject: Re: software to control Olympus IX81 for video microscopy > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Check out MicroManager when you get a chance. > > John Oreopoulos > > On 2011-12-28, at 2:19 PM, Bernhard Kuhn <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Dear All, > > > > We have an Olympus IX81 with a motorized stage Prior Proscan II, > microscopy > > control unit IX2-USB and Olympus Laser- Based Autofocus System (IX2-ZDC). > > > > The camera is a Hamamatsu EM-CCD C9100-02, controlled with A3472-07. > > > > We are using this unit for multi-location, multi-time microscopy ~30 > > locations for ~150 acquisitions per experiment. > > > > What software and operating system do people recommend to control this > > setup? > > > > Is there anyone with a similar setup who can make software > recommendations? > > > > Thank you and Happy New Year, > > > > Bernhard Kuhn > -- ----------------------------------------------------------- Johan Henriksson PhD student, Karolinska Institutet http://mahogny.areta.org http://www.endrov.net |
|
tineke vendrig |
|
|
In reply to this post by Paul Rigby-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Daer all, We are working with Andor software and this will work fine, after having some additional back-up/support. Best, Tineke Vendrig technical engineer optical microscopy Delft University of Technology Bionano Science Kavli Institute of Nanoscience Lorentzweg 1 2628LJ Delft room F185 Tel: +31 15 2781428 Fax:+31 15 2781202 email: [hidden email] mobile phone: +31 624341412 2012/1/5 Paul Rigby <[hidden email]> > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi All, > I have also been trying to control an IX81 with almost the identical > hardware to Bernhard (Prior stage, ZDC etc) using Micromanager. However, I > have been unable to get our computer (and Micromanager) to successfully > talk with the Olympus control box via the serial port. I suspect the serial > port settings. If anyone has done this successfully, can you please let > me/us know what settings you used or what has worked best. > > Many thanks. > Paul > > Assoc. Prof. Paul Rigby > Centre for Microscopy, Characterisation & Analysis (M510) > The University of Western Australia > 35 Stirling Highway > Crawley WA 6007 > Australia > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of John Oreopoulos > Sent: Thursday, 29 December 2011 4:37 AM > To: [hidden email] > Subject: Re: software to control Olympus IX81 for video microscopy > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Check out MicroManager when you get a chance. > > John Oreopoulos > > On 2011-12-28, at 2:19 PM, Bernhard Kuhn <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Dear All, > > > > We have an Olympus IX81 with a motorized stage Prior Proscan II, > microscopy > > control unit IX2-USB and Olympus Laser- Based Autofocus System (IX2-ZDC). > > > > The camera is a Hamamatsu EM-CCD C9100-02, controlled with A3472-07. > > > > We are using this unit for multi-location, multi-time microscopy ~30 > > locations for ~150 acquisitions per experiment. > > > > What software and operating system do people recommend to control this > > setup? > > > > Is there anyone with a similar setup who can make software > recommendations? > > > > Thank you and Happy New Year, > > > > Bernhard Kuhn > |
|
Nico Stuurman |
|
|
In reply to this post by Paul Rigby-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Paul, > I have also been trying to control an IX81 with almost the identical hardware to Bernhard (Prior stage, ZDC etc) using Micromanager. However, I have been unable to get our computer (and Micromanager) to successfully talk with the Olympus control box via the serial port. I suspect the serial port settings. If anyone has done this successfully, can you please let me/us know what settings you used or what has worked best. It can be challenging to get hardware to cooperate! However, there are many labs that are using Micro-Manager to operate an Olympus IX81. If you still have trouble after increasing the serial port timeout as suggested by Johan, please use the menu entry "Help"->"Report a Problem". This will give me a chance to figure out what is going on and help you make it work. Also, the Micro-Manager mailing list ( https://lists.sourceforge.net/lists/listinfo/micro-manager-general ) may be a more useful forum for issues such as these. Best, Nico |