HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello,
This is not strictly a confocal question, but reasonably relevant and it's not been recently discussed from what I can tell from the archives.
I was hoping to find out at how many hours people change the HBO/Mercury short arc lamps (specifically Osram 103W). The microscope manufacturers (Zeiss & Leica) always recommend 300 hours, due to the potential dangers of explosion when using them for longer... but what do users/facilities tend to do?
Someone recently told me that the lamps are now made in such a way that explosions are less common that ~10 years ago (when discussion on this list regarding explosions were more common)
Thanks in advance,
Graham
 
-----
Dr Graham Wright
Microscopy & Imaging Facility
 
Temasek Life Sciences Laboratory
1 Research Link
National University of Singapore
Singapore 117604
 
P:    +65 6872 8406
M:   +65 8256 7916
E:     [hidden email]
W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm
Michael Weber-4 Michael Weber-4
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Graham,

I tend to change the normal HBO bulbs after 300-400 hours. The
recommendation on our setups says 200 hours. Luckily the number of
HBO-installations is constantly decreasing!

Michael


Graham Wright wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> Hello,
> This is not strictly a confocal question, but reasonably relevant and
> it's not been recently discussed from what I can tell from the archives.
> I was hoping to find out at how many hours people change the HBO/Mercury
> short arc lamps (specifically Osram 103W). The microscope manufacturers
> (Zeiss & Leica) always recommend 300 hours, due to the potential dangers
> of explosion when using them for longer... but what do users/facilities
> tend to do?
> Someone recently told me that the lamps are now made in such a way that
> explosions are less common that ~10 years ago (when discussion on this
> list regarding explosions were more common)
> Thanks in advance,
> Graham
>  
> -----
> Dr Graham Wright
> Microscopy & Imaging Facility
>  
> Temasek Life Sciences Laboratory
> 1 Research Link
> National University of Singapore
> Singapore 117604
>  
> P:    +65 6872 8406
> M:   +65 8256 7916
> E:     [hidden email] <mailto:[hidden email]>
> W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm
> <http://www.bioinformatics.tll.org.sg/labs/microscopy/index.htm>
Nowell, Cameron Nowell, Cameron
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Graham Wright-5
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Graham,
                We get our lamps (HBO 103/W2) out to around 600 hours. Our practice is to use new ones in our scopes that take images for analysis (widefield uprights and live imaging scopes) we then change them when they hit 300 hours with new ones. The ones that have done 300 hours we use in our confocals as they only need the light to find a sample prior to confocal imaging. These lamps then usually stop firing up after 600 hours (our record is 850 hours)
 
For us the problem is not exploding lamps but greatly reduced light intensity. If you compare a lamp that has done 600 hours to one that is brand new, there is a great deal of difference.
 
We have been doing this for a year and a half now and have not had any problems. When we first started to do it, we asked a lot of people about the exploding lamps. No one we spoke to had had it happen to them but everyone new someone (or new someone who new someone) who had had it happen in the past.
 
 
Cheers
 
 
Cam
 
 
 
Cameron Nowell
Microscopy Research and Imaging Facility
Cell Cycle and Development
Peter MacCallum Cancer Centre
7 St Andrews Place
East Melbourne, 3002
Victoria AUSTRALIA

Phone: +61396563759
Fax: +61396561411
Mobile: +61422882700
 

________________________________

From: Confocal Microscopy List on behalf of Graham Wright
Sent: Fri 25/04/2008 11:30 AM
To: [hidden email]
Subject: HBO/Mercury lamp lifetimes


Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 
Hello,
This is not strictly a confocal question, but reasonably relevant and it's not been recently discussed from what I can tell from the archives.
I was hoping to find out at how many hours people change the HBO/Mercury short arc lamps (specifically Osram 103W). The microscope manufacturers (Zeiss & Leica) always recommend 300 hours, due to the potential dangers of explosion when using them for longer... but what do users/facilities tend to do?
Someone recently told me that the lamps are now made in such a way that explosions are less common that ~10 years ago (when discussion on this list regarding explosions were more common)
Thanks in advance,
Graham
 
-----
Dr Graham Wright
Microscopy & Imaging Facility
 
Temasek Life Sciences Laboratory
1 Research Link
National University of Singapore
Singapore 117604
 
P:    +65 6872 8406
M:   +65 8256 7916
E:     [hidden email]
W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm


This email (including any attachments) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee.  If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email.  Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.
Hege Avsnes Dale Hege Avsnes Dale
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Graham Wright-5
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi.
We also normally change after approx. 600 h on our confocals, but have had them going for even longer without any explosions... Just decreasing intensity as mentioned here before.
Hege


Graham Wright wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello,
This is not strictly a confocal question, but reasonably relevant and it's not been recently discussed from what I can tell from the archives.
I was hoping to find out at how many hours people change the HBO/Mercury short arc lamps (specifically Osram 103W). The microscope manufacturers (Zeiss & Leica) always recommend 300 hours, due to the potential dangers of explosion when using them for longer... but what do users/facilities tend to do?
Someone recently told me that the lamps are now made in such a way that explosions are less common that ~10 years ago (when discussion on this list regarding explosions were more common)
Thanks in advance,
Graham
 
-----
Dr Graham Wright
Microscopy & Imaging Facility
 
Temasek Life Sciences Laboratory
1 Research Link
National University of Singapore
Singapore 117604
 
P:    +65 6872 8406
M:   +65 8256 7916
E:     [hidden email]
W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm

Dries Vercauteren Dries Vercauteren
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all,

regarding this topic, I'd like to bring up an current issue in our lab;
we're using the X-Cite® 120 metal halid lamp from XFO. http://www.exfo-lifesciences.com/x-cite/prod-120.asp
A couple of weeks ago, I remembered to have this lamp on 760 burning hours, but last week I suddenly noticed that this lamp displayed 3600 burning hours! Which is impossible.. The lamp is garantueed for 2000 hrs, but does anyone of you know how this would be possible, or what the risk is to still use this lamp (intensity is still high), as far as I know, there is net really an explosion danger with these metal halide lamps, right?

Thanks in advance!

Dries.

2008/4/25 Hege Avsnes Dale <[hidden email]>:
Hi.
We also normally change after approx. 600 h on our confocals, but have had them going for even longer without any explosions... Just decreasing intensity as mentioned here before.
Hege


Graham Wright wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello,
This is not strictly a confocal question, but reasonably relevant and it's not been recently discussed from what I can tell from the archives.
I was hoping to find out at how many hours people change the HBO/Mercury short arc lamps (specifically Osram 103W). The microscope manufacturers (Zeiss & Leica) always recommend 300 hours, due to the potential dangers of explosion when using them for longer... but what do users/facilities tend to do?
Someone recently told me that the lamps are now made in such a way that explosions are less common that ~10 years ago (when discussion on this list regarding explosions were more common)
Thanks in advance,
Graham
 
-----
Dr Graham Wright
Microscopy & Imaging Facility
 
Temasek Life Sciences Laboratory
1 Research Link
National University of Singapore
Singapore 117604
 
P:    +65 6872 8406
M:   +65 8256 7916
E:     [hidden email]
W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm




--
Dries Vercauteren, PhD student
Master of Bioscience Engineering: Cell and Gene Biotechnology

Ghent Research Group on Nanomedicines
www.ugent.be/fw/en/research/biofys
Faculty of pharmaceutical sciences, Ghent University
Harelbekestraat 72, 9000 Ghent
Belgium

Phone: +329/264 80 49
Mobile: +32485/30 69 80
E-mail: [hidden email]
[hidden email]
Michael Herron Michael Herron
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Graham,

I have not seen that glitch before, but last winter I got a new bulb from them that registered 400 hours when I installed it.  I am guessing that the little chips they use to track the time are a bit buggy.  It is interesting that you bulb continues to run even though it is reading 2000 hours.  That might indicate that the counter that normally turns these bulbs off when they reach the end of their life is only looking for that number to flip and after that it does not check. I am am not certain the bulb will explode, but it is my understanding that they will eventually.

Mike

On Apr 25, 2008, at 5:15 AM, Dries Vercauteren wrote:

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all,

regarding this topic, I'd like to bring up an current issue in our lab;
we're using the X-Cite® 120 metal halid lamp from XFO. http://www.exfo-lifesciences.com/x-cite/prod-120.asp
A couple of weeks ago, I remembered to have this lamp on 760 burning hours, but last week I suddenly noticed that this lamp displayed 3600 burning hours! Which is impossible.. The lamp is garantueed for 2000 hrs, but does anyone of you know how this would be possible, or what the risk is to still use this lamp (intensity is still high), as far as I know, there is net really an explosion danger with these metal halide lamps, right?

Thanks in advance!

Dries.

2008/4/25 Hege Avsnes Dale <[hidden email]>:
Hi.
We also normally change after approx. 600 h on our confocals, but have had them going for even longer without any explosions... Just decreasing intensity as mentioned here before.
Hege


Graham Wright wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello,
This is not strictly a confocal question, but reasonably relevant and it's not been recently discussed from what I can tell from the archives.
I was hoping to find out at how many hours people change the HBO/Mercury short arc lamps (specifically Osram 103W). The microscope manufacturers (Zeiss & Leica) always recommend 300 hours, due to the potential dangers of explosion when using them for longer... but what do users/facilities tend to do?
Someone recently told me that the lamps are now made in such a way that explosions are less common that ~10 years ago (when discussion on this list regarding explosions were more common)
Thanks in advance,
Graham
 
-----
Dr Graham Wright
Microscopy & Imaging Facility
 
Temasek Life Sciences Laboratory
1 Research Link
National University of Singapore
Singapore 117604
 
P:    +65 6872 8406
M:   +65 8256 7916
E:     [hidden email]
W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm




--
Dries Vercauteren, PhD student
Master of Bioscience Engineering: Cell and Gene Biotechnology

Ghent Research Group on Nanomedicines
www.ugent.be/fw/en/research/biofys
Faculty of pharmaceutical sciences, Ghent University
Harelbekestraat 72, 9000 Ghent
Belgium

Phone: +329/264 80 49
Mobile: +32485/30 69 80
E-mail: [hidden email]
[hidden email]




Michael J. Herron,  U of MN, Dept. of Entomology

  [hidden email]

     612-624-3688 (office) 612-625-5299 (FAX)




Aryeh Weiss Aryeh Weiss
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

My first EXFO lamp lasted to 3300 hours. I do long runs with relatively few
on/off cycles, so I expected the lamp to live past the rated 1500 hours.
However, my second lamp spontaneously turned off during overnight runs once it
got past 800 hours. This was obviously unacceptable, so we replaced it. The
replacement did something similar after around 700 hours, but it did not happen
as often. Still, one does not want even a small chance of a long experiment
failing because the lamp turned off. We checked the air filter, but that was
clean, so we have no idea what this is about. (I filed a failure report, and
that was the last I heard about it.)

I would be interested in knowing whether others on this list have had premature
EXFO lamp failures.

--aryeh
--
Aryeh Weiss
School of Engineering
Bar Ilan University
Ramat Gan 52900 Israel

Ph:  972-3-5317638
FAX: 972-3-7384050

Michael Herron wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Graham,
>
> I have not seen that glitch before, but last winter I got a new bulb
> from them that registered 400 hours when I installed it.  I am guessing
> that the little chips they use to track the time are a bit buggy.  It is
> interesting that you bulb continues to run even though it is reading
> 2000 hours.  That might indicate that the counter that normally turns
> these bulbs off when they reach the end of their life is only looking
> for that number to flip and after that it does not check. I am am not
> certain the bulb will explode, but it is my understanding that they will
> eventually.
>
> Mike
>
> On Apr 25, 2008, at 5:15 AM, Dries Vercauteren wrote:
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all,
>>
>> regarding this topic, I'd like to bring up an current issue in our lab;
>> we're using the X-Cite® 120 metal halid lamp from XFO.
>> http://www.exfo-lifesciences.com/x-cite/prod-120.asp
>> A couple of weeks ago, I remembered to have this lamp on 760 burning
>> hours, but last week I suddenly noticed that this lamp displayed 3600
>> burning hours! Which is impossible.. The lamp is garantueed for 2000
>> hrs, but does anyone of you know how this would be possible, or what
>> the risk is to still use this lamp (intensity is still high), as far
>> as I know, there is net really an explosion danger with these metal
>> halide lamps, right?
>>
>> Thanks in advance!
>>
>> Dries.
>>
>> 2008/4/25 Hege Avsnes Dale <[hidden email]
>> <mailto:[hidden email]>>:
>>
>>     Search the CONFOCAL archive at
>>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>     Hi.
>>     We also normally change after approx. 600 h on our confocals, but
>>     have had them going for even longer without any explosions... Just
>>     decreasing intensity as mentioned here before.
>>     Hege
>>
>>
>>     Graham Wright wrote:
>>>     Search the CONFOCAL archive at
>>>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>>     Hello,
>>>     This is not strictly a confocal question, but reasonably relevant
>>>     and it's not been recently discussed from what I can tell from
>>>     the archives.
>>>     I was hoping to find out at how many hours people change the
>>>     HBO/Mercury short arc lamps (specifically Osram 103W). The
>>>     microscope manufacturers (Zeiss & Leica) always recommend 300
>>>     hours, due to the potential dangers of explosion when using them
>>>     for longer... but what do users/facilities tend to do?
>>>     Someone recently told me that the lamps are now made in such a
>>>     way that explosions are less common that ~10 years ago (when
>>>     discussion on this list regarding explosions were more common)
>>>     Thanks in advance,
>>>     Graham
>>>      
>>>     -----
>>>     Dr Graham Wright
>>>     Microscopy & Imaging Facility
>>>      
>>>     Temasek Life Sciences Laboratory
>>>     1 Research Link
>>>     National University of Singapore
>>>     Singapore 117604
>>>      
>>>     P:    +65 6872 8406
>>>     M:   +65 8256 7916
>>>     E:     [hidden email] <mailto:[hidden email]>
>>>     W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm
>>>     <http://www.bioinformatics.tll.org.sg/labs/microscopy/index.htm>
>>
>>
>>
>>
>> --
>> Dries Vercauteren, PhD student
>> Master of Bioscience Engineering: Cell and Gene Biotechnology
>>
>> Ghent Research Group on Nanomedicines
>> www.ugent.be/fw/en/research/biofys
>> <http://www.ugent.be/fw/en/research/biofys>
>> Faculty of pharmaceutical sciences, Ghent University
>> Harelbekestraat 72, 9000 Ghent
>> Belgium
>>
>> Phone: +329/264 80 49
>> Mobile: +32485/30 69 80
>> E-mail: [hidden email] <mailto:[hidden email]>
>> [hidden email] <mailto:[hidden email]>
>
>
>
>
> Michael J. Herron,  U of MN, Dept. of Entomology
>
>   [hidden email] <mailto:[hidden email]>
>
>      612-624-3688 (office) 612-625-5299 (FAX)
>
>
>
>
Petrak, Lara J. Petrak, Lara J.
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Aryeh,  

I work in a multi-user facility, and we have several EXFO lamps.  Many
of the lamps last through their expected lifetimes (either 1200 or 2000
hours, depending on which model lamp housing they're in).  However, we
have had quite a few that failed very early - as early at 500 hours in
some cases.  Interestingly, the really early failures tend to occur on
systems that are used for multi-day experiments - that is, the lamps are
on continuously for days at a time, with few on/off cycles, as you say.
The good news is that the company has replaced many of these lamps after
we filed failure reports.  You do have to be persistent, and the process
of getting a replacement takes a little while.

I too would be interested in hearing about other EXFO users experiences!

And in reference to Graham's original question, we use Ushio USH-102DH
arc lamps in our standard mercury lamp housings.  They are rated for 200
hours, so we always replace them at that point.

Best,
Lara

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Aryeh Weiss
Sent: Friday, April 25, 2008 8:46 AM
To: [hidden email]
Subject: Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

My first EXFO lamp lasted to 3300 hours. I do long runs with relatively
few
on/off cycles, so I expected the lamp to live past the rated 1500 hours.

However, my second lamp spontaneously turned off during overnight runs
once it
got past 800 hours. This was obviously unacceptable, so we replaced it.
The
replacement did something similar after around 700 hours, but it did not
happen
as often. Still, one does not want even a small chance of a long
experiment
failing because the lamp turned off. We checked the air filter, but that
was
clean, so we have no idea what this is about. (I filed a failure report,
and
that was the last I heard about it.)

I would be interested in knowing whether others on this list have had
premature
EXFO lamp failures.

--aryeh
--
Aryeh Weiss
School of Engineering
Bar Ilan University
Ramat Gan 52900 Israel

Ph:  972-3-5317638
FAX: 972-3-7384050

Michael Herron wrote:
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Graham,
>
> I have not seen that glitch before, but last winter I got a new bulb
> from them that registered 400 hours when I installed it.  I am
guessing
> that the little chips they use to track the time are a bit buggy.  It
is
> interesting that you bulb continues to run even though it is reading
> 2000 hours.  That might indicate that the counter that normally turns
> these bulbs off when they reach the end of their life is only looking
> for that number to flip and after that it does not check. I am am not
> certain the bulb will explode, but it is my understanding that they
will

> eventually.
>
> Mike
>
> On Apr 25, 2008, at 5:15 AM, Dries Vercauteren wrote:
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all,
>>
>> regarding this topic, I'd like to bring up an current issue in our
lab;
>> we're using the X-Cite(r) 120 metal halid lamp from XFO.
>> http://www.exfo-lifesciences.com/x-cite/prod-120.asp
>> A couple of weeks ago, I remembered to have this lamp on 760 burning
>> hours, but last week I suddenly noticed that this lamp displayed 3600

>> burning hours! Which is impossible.. The lamp is garantueed for 2000
>> hrs, but does anyone of you know how this would be possible, or what
>> the risk is to still use this lamp (intensity is still high), as far
>> as I know, there is net really an explosion danger with these metal
>> halide lamps, right?
>>
>> Thanks in advance!
>>
>> Dries.
>>
>> 2008/4/25 Hege Avsnes Dale <[hidden email]
>> <mailto:[hidden email]>>:
>>
>>     Search the CONFOCAL archive at
>>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>     Hi.
>>     We also normally change after approx. 600 h on our confocals, but
>>     have had them going for even longer without any explosions...
Just
>>     decreasing intensity as mentioned here before.
>>     Hege
>>
>>
>>     Graham Wright wrote:
>>>     Search the CONFOCAL archive at
>>>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>>     Hello,
>>>     This is not strictly a confocal question, but reasonably
relevant

>>>     and it's not been recently discussed from what I can tell from
>>>     the archives.
>>>     I was hoping to find out at how many hours people change the
>>>     HBO/Mercury short arc lamps (specifically Osram 103W). The
>>>     microscope manufacturers (Zeiss & Leica) always recommend 300
>>>     hours, due to the potential dangers of explosion when using them
>>>     for longer... but what do users/facilities tend to do?
>>>     Someone recently told me that the lamps are now made in such a
>>>     way that explosions are less common that ~10 years ago (when
>>>     discussion on this list regarding explosions were more common)
>>>     Thanks in advance,
>>>     Graham
>>>      
>>>     -----
>>>     Dr Graham Wright
>>>     Microscopy & Imaging Facility
>>>      
>>>     Temasek Life Sciences Laboratory
>>>     1 Research Link
>>>     National University of Singapore
>>>     Singapore 117604
>>>      
>>>     P:    +65 6872 8406
>>>     M:   +65 8256 7916
>>>     E:     [hidden email] <mailto:[hidden email]>
>>>     W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm
>>>     <http://www.bioinformatics.tll.org.sg/labs/microscopy/index.htm>
>>
>>
>>
>>
>> --
>> Dries Vercauteren, PhD student
>> Master of Bioscience Engineering: Cell and Gene Biotechnology
>>
>> Ghent Research Group on Nanomedicines
>> www.ugent.be/fw/en/research/biofys
>> <http://www.ugent.be/fw/en/research/biofys>
>> Faculty of pharmaceutical sciences, Ghent University
>> Harelbekestraat 72, 9000 Ghent
>> Belgium
>>
>> Phone: +329/264 80 49
>> Mobile: +32485/30 69 80
>> E-mail: [hidden email]
<mailto:[hidden email]>

>> [hidden email] <mailto:[hidden email]>
>
>
>
>
> Michael J. Herron,  U of MN, Dept. of Entomology
>
>   [hidden email] <mailto:[hidden email]>
>
>      612-624-3688 (office) 612-625-5299 (FAX)
>
>
>
>
George Ring George Ring
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Graham Wright-5
We change our bulb between 300 and 400 hours.  We had a mercury bulb explode once several years ago,  and it caused at least a couple  thousand dollars damage to the microscope (condenser lens, housing, etc).  To me, the savings of $50/yr by letting the bulb go longer isn't worth the potential cost in money and downtime of an explosion.
...just my 2 cents worth.
 
George
 
George Ring, Ph.D.
Dept. of Cell and Developmental Biology
SUNY Upstate Medical University
750 E. Adams St.
Syracuse NY  13210
Tel. (315) 464-8595
FAX (315) 464-8535
email: [hidden email]

>>> Graham Wright <[hidden email]> 4/24/08 9:30:11 PM >>>
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello,
This is not strictly a confocal question, but reasonably relevant and it's not been recently discussed from what I can tell from the archives.
I was hoping to find out at how many hours people change the HBO/Mercury short arc lamps (specifically Osram 103W). The microscope manufacturers (Zeiss & Leica) always recommend 300 hours, due to the potential dangers of explosion when using them for longer... but what do users/facilities tend to do?
Someone recently told me that the lamps are now made in such a way that explosions are less common that ~10 years ago (when discussion on this list regarding explosions were more common)
Thanks in advance,
Graham
 
-----
Dr Graham Wright
Microscopy & Imaging Facility
 
Temasek Life Sciences Laboratory
1 Research Link
National University of Singapore
Singapore 117604
 
P:    +65 6872 8406
M:   +65 8256 7916
E:     [hidden email]
W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm
Barbara Foster Barbara Foster
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Dries Vercauteren
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear Dries,

I'm sure that someone from XFO can comment more accurately, but I helped launch that product some years ago.  Here are some of the differences:
a. The bulb is like the bulb you have in a slide projector rather than an arc source.  No risk of explosion
b. The rated lifetimes that I knew about were on the ordr of 3000 hrs (compared to 200 for HBO)
c. They also do not darken with age (like W filament sources)

Hope this is helpful

Barbara Foster, President

Microscopy/Microscopy Education
7101 Royal Glen Trail, Suite A
McKinney TX 75070
P: (972)924-5310
Skype: fostermme
W: www.MicroscopyEducation.com


MME is now scheduling customized, on-site courses through Sept 2008.  Call us today for details.





t 10:28 PM 4/24/2008, you wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all,

regarding this topic, I'd like to bring up an current issue in our lab;
we're using the X-Cite® 120 metal halid lamp from XFO. http://www.exfo-lifesciences.com/x-cite/prod-120.asp
A couple of weeks ago, I remembered to have this lamp on 760 burning hours, but last week I suddenly noticed that this lamp displayed 3600 burning hours! Which is impossible.. The lamp is garantueed for 2000 hrs, but does anyone of you know how this would be possible, or what the risk is to still use this lamp (intensity is still high), as far as I know, there is net really an explosion danger with these metal halide lamps, right?

Thanks in advance!

Dries.

2008/4/25 Hege Avsnes Dale <[hidden email] >:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hi.
We also normally change after approx. 600 h on our confocals, but have had them going for even longer without any explosions... Just decreasing intensity as mentioned here before.
Hege
Graham Wright wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello,
This is not strictly a confocal question, but reasonably relevant and it's not been recently discussed from what I can tell from the archives.
I was hoping to find out at how many hours people change the HBO/Mercury short arc lamps (specifically Osram 103W). The microscope manufacturers (Zeiss & Leica) always recommend 300 hours, due to the potential dangers of explosion when using them for longer... but what do users/facilities tend to do?
Someone recently told me that the lamps are now made in such a way that explosions are less common that ~10 years ago (when discussion on this list regarding explosions were more common)
Thanks in advance,
Graham
 
-----
Dr Graham Wright
Microscopy & Imaging Facility
 
Temasek Life Sciences Laboratory
1 Research Link
National University of Singapore
Singapore 117604
 
P:    +65 6872 8406
M:   +65 8256 7916
E:     [hidden email]
W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm



--
Dries Vercauteren, PhD student
Master of Bioscience Engineering: Cell and Gene Biotechnology

Ghent Research Group on Nanomedicines
www.ugent.be/fw/en/research/biofys
Faculty of pharmaceutical sciences, Ghent University
Harelbekestraat 72, 9000 Ghent
Belgium

Phone: +329/264 80 49
Mobile: +32485/30 69 80
E-mail: [hidden email]
[hidden email]
Boisvert, Karen Boisvert, Karen
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Graham Wright-5
I use Osram hbo 50W/AC  mercury lamps/bulbs.
 
In my experience - 2 explosions later - the hours are meaningless - it is the number of ignitions that need to be counted!  1 ignition/day. I change the bulb at about 50 ignitions.
 
When I inherited the Confocal Lab a log book for the lab did not exist.  The mercury bulb meter reading is now recorded prior to ignition and when the bulb is turned off. I had a bulb explode after 71 hours but now that I keep track of the number of ignitions they have lasted up to >550 hours!
 
Feel free to call me, Karen
 
Karen Dalecki Boisvert
Division of Comparative Pathology
Confocal Microscopy Unit
Harvard University
TEL:   (508) 624-8013
FAX:  (508) 624-8181
email:  [hidden email] 
-----Original Message-----
From: Confocal Microscopy List on behalf of Graham Wright
Sent: Thu 4/24/2008 9:30 PM
To: [hidden email]
Cc:
Subject: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello,
This is not strictly a confocal question, but reasonably relevant and it's not been recently discussed from what I can tell from the archives.
I was hoping to find out at how many hours people change the HBO/Mercury short arc lamps (specifically Osram 103W). The microscope manufacturers (Zeiss & Leica) always recommend 300 hours, due to the potential dangers of explosion when using them for longer... but what do users/facilities tend to do?
Someone recently told me that the lamps are now made in such a way that explosions are less common that ~10 years ago (when discussion on this list regarding explosions were more common)
Thanks in advance,
Graham
 
-----
Dr Graham Wright
Microscopy & Imaging Facility
 
Temasek Life Sciences Laboratory
1 Research Link
National University of Singapore
Singapore 117604
 
P:    +65 6872 8406
M:   +65 8256 7916
E:     [hidden email]
W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm
McDonald, David L McDonald, David L
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Graham Wright-5
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
We change all our Hg lamps at the recommended time including the HBO 103W in our Nikon E800.  It seems several years ago Osram was having some QC issues, we had three 103W lamps in a row fail before reaching 300 hours.  One only lasted 9 hours!  In the last year we've also had two 100W lamps fail before reaching 50 hours.

Dave

At 06:30 PM 4/24/2008, you wrote:
Hello,
This is not strictly a confocal question, but reasonably relevant and it's not been recently discussed from what I can tell from the archives.
I was hoping to find out at how many hours people change the HBO/Mercury short arc lamps (specifically Osram 103W). The microscope manufacturers (Zeiss & Leica) always recommend 300 hours, due to the potential dangers of explosion when using them for longer... but what do users/facilities tend to do?
Someone recently told me that the lamps are now made in such a way that explosions are less common that ~10 years ago (when discussion on this list regarding explosions were more common)
Thanks in advance,
Graham

Dave McDonald
Scientific Imaging Lab
Fred Hutchinson Cancer Research Center
1100 Fairview Avenue North, DE-512
Seattle, WA 98109
206-667-4205
http://www.fhcrc.org

Dale Callaham Dale Callaham
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Graham Wright-5
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Graham,

Old equipment (pre-80?) tended to control the current via an AC ballast
(even if the current was subsequently rectified for a DC lamp..). This
is also true of the inexpensive HBO 50 AC lamp units. If done this way,
using inductive reactance, there was not much of a limit to the voltage
that could be supplied and as the electrodes erode to wider openings the
voltage increases while the current is "regulated"; the result is much
higher power dissipation and that is when explosions would occur if the
lamp did not get changed. If left that long, there also tended to be
oxidation on the connections that added to the power dissipation.

I think most modern gear regulates the current actively, but has some
control/limit to the compliance voltage and thus there is a limit to the
power dissipation. I haven't heard of an exploding HBO 100 or 103 used
with modern equipment. I'm sure that with modern gear failures can
happen, but with the old gear it could be expected if the lamp was not
changed.

Dale




Graham Wright wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> Hello,
> This is not strictly a confocal question, but reasonably relevant and
> it's not been recently discussed from what I can tell from the archives.
> I was hoping to find out at how many hours people change the HBO/Mercury
> short arc lamps (specifically Osram 103W). The microscope manufacturers
> (Zeiss & Leica) always recommend 300 hours, due to the potential dangers
> of explosion when using them for longer... but what do users/facilities
> tend to do?
> Someone recently told me that the lamps are now made in such a way that
> explosions are less common that ~10 years ago (when discussion on this
> list regarding explosions were more common)
> Thanks in advance,
> Graham
>  
> -----
> Dr Graham Wright
> Microscopy & Imaging Facility
>  
> Temasek Life Sciences Laboratory
> 1 Research Link
> National University of Singapore
> Singapore 117604
>  
> P:    +65 6872 8406
> M:   +65 8256 7916
> E:     [hidden email] <mailto:[hidden email]>
> W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm
> <http://www.bioinformatics.tll.org.sg/labs/microscopy/index.htm>
Michael Weber-4 Michael Weber-4
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dale,

we had two explosions with HBO 103W/2 over the last year, so it's not a
matter of the bulb. However, it turned out that the power supplies where
faulty, so I agree with you at this point.

-----

Regarding the Exfo system, we had two systems with faulty counter as
well - apparently a known problem. So far we have better experiences
with the Prior Lumen 200 light source.

Michael


-----


Dale Callaham wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi Graham,
>
> Old equipment (pre-80?) tended to control the current via an AC ballast
> (even if the current was subsequently rectified for a DC lamp..). This
> is also true of the inexpensive HBO 50 AC lamp units. If done this way,
> using inductive reactance, there was not much of a limit to the voltage
> that could be supplied and as the electrodes erode to wider openings the
> voltage increases while the current is "regulated"; the result is much
> higher power dissipation and that is when explosions would occur if the
> lamp did not get changed. If left that long, there also tended to be
> oxidation on the connections that added to the power dissipation.
>
> I think most modern gear regulates the current actively, but has some
> control/limit to the compliance voltage and thus there is a limit to the
> power dissipation. I haven't heard of an exploding HBO 100 or 103 used
> with modern equipment. I'm sure that with modern gear failures can
> happen, but with the old gear it could be expected if the lamp was not
> changed.
>
> Dale
>
>
>
>
> Graham Wright wrote:
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>> Hello,
>> This is not strictly a confocal question, but reasonably relevant and
>> it's not been recently discussed from what I can tell from the archives.
>> I was hoping to find out at how many hours people change the
>> HBO/Mercury short arc lamps (specifically Osram 103W). The microscope
>> manufacturers (Zeiss & Leica) always recommend 300 hours, due to the
>> potential dangers of explosion when using them for longer... but what
>> do users/facilities tend to do?
>> Someone recently told me that the lamps are now made in such a way
>> that explosions are less common that ~10 years ago (when discussion on
>> this list regarding explosions were more common)
>> Thanks in advance,
>> Graham
>>  
>> -----
>> Dr Graham Wright
>> Microscopy & Imaging Facility
>>  
>> Temasek Life Sciences Laboratory
>> 1 Research Link
>> National University of Singapore
>> Singapore 117604
>>  
>> P:    +65 6872 8406
>> M:   +65 8256 7916
>> E:     [hidden email] <mailto:[hidden email]>
>> W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm
>> <http://www.bioinformatics.tll.org.sg/labs/microscopy/index.htm>
Russell Spear Russell Spear
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Boisvert, Karen
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I've been using HBO 100 W s, for about 30 years, and have had 2
explosions in that time, one long ago and one approx. 3 yrs past.  Both
could be attributed to a high number of on off cycles (something I
discourage).  I nomally cycle a bulb no more than twice a day, and need
to relamp about every 40-45 days in a busy lab with 3 fluorescent
scopes.  We always replace at 200 hrs. the cost of the bulb is small
compared to the damage to the equipment and time and cost of
decontaminating the microscope rooms.

Russ



Russell N. Spear
Sr. Research Specialist
Dept. of Plant Pathology
Univ. of Wisconsin
1630 Linden Dr.
Madison WI 53706

voice 608.263.2093
fax     608.263.2626
Jon Ekman Jon Ekman
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Graham Wright-5
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

One of our SP2 confocals has a HBO 103/W2, the last time I changed it, the timer read 633hrs. We also have a 50W/AC/L1 bulb on our older SP2, the power supply timer says nearly 13,000hrs. On both confocals, we change the bulbs when people complain that they are too dim, this ends up being about 3 times a year, and anywhere from 200hrs-600hrs.

 

As for X-cite lamps, we have models XI120 & XI120-XL, the lamps in these last about 1400hrs. We did have to replace the ballast on one (the XI120) after a series of low lamp hr failures, first at 500hrs, then at 300hrs, then no strike on a fresh lamp. The EXFO people fixed our ballast with a 1 week turnaround, and the lamps that had 300hrs and 500hrs were either still good (at 300hours) or were replaced by EXFO for free. We also had the liquid light guide fail at about 2600hrs. A user noticed the power drop and I did some quick tests and found the power at the objective dropped by 90% when compared against the liquid light guide on our second X-cite system. Again, the replacement part arrived from EXFO within a couple days.

 

Our widefield fluorescent scopes have the X-cite lamps and they both get used quite heavily. There are days (sometimes whole weeks) when the lamp is off for only 6hours.

 

I do have an AttoArc Hg 100W system that has had its ballast and various chips on it main board replaced over the past few years. I watch this system carefully and change the bulb on it like clockwork at about 200hrs, mainly because it has been serviced so much.

 

Cheers,

 

Jon

 

Jonathan M. Ekman

Imaging Technology Group

Beckman Institute for Advanced Science and Technology University of Illinois at Urbana-Champaign

405 N. Mathews Avenue

Urbana, IL 61801 USA

Tel: 217-244-6292

Fax: 217-244-6219


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Graham Wright
Sent: Thursday, April 24, 2008 8:30 PM
To: [hidden email]
Subject: HBO/Mercury lamp lifetimes

 

Hello,

This is not strictly a confocal question, but reasonably relevant and it's not been recently discussed from what I can tell from the archives.

I was hoping to find out at how many hours people change the HBO/Mercury short arc lamps (specifically Osram 103W). The microscope manufacturers (Zeiss & Leica) always recommend 300 hours, due to the potential dangers of explosion when using them for longer... but what do users/facilities tend to do?

Someone recently told me that the lamps are now made in such a way that explosions are less common that ~10 years ago (when discussion on this list regarding explosions were more common)

Thanks in advance,

Graham

 

-----
Dr Graham Wright
Microscopy & Imaging Facility

 

Temasek Life Sciences Laboratory
1 Research Link
National University of Singapore
Singapore 117604

 

P:    +65 6872 8406
M:   +65 8256 7916
E:     [hidden email]
W:    www.bioinformatics.tll.org.sg/labs/microscopy/index.htm

Ignatius, Mike Ignatius, Mike
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Russell Spear
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

An important side bar in this discussion, is what warning signs are
there that the bulb is too old and possibly going to fail/explode.  

For me, the earliest signs that the arc gap has grown to impossible
widths were difficulty in starting the bulb to begin with, flickering
when on, long time to stabilize etc.  

The more dreadful indication is an electrical burning smell when it is
on and about to blow.

In my one experience it smelled like insulation burning off of wiring.
Apparently the bulb super heats before failure?  I would be curious to
hear if others have had this warning sign before the bulb blew.  I
stupidly got up, and sniffed around the scope to see what the source was
and wham the bulb blew in my face.  I thanked Nikon for their sturdy
housing, no glass came through, though the diffuser lens inside did
shatter, and the mercury vapor surely escaped.  After it has blown,
always good to leave the area for a while to let the gas escape.  

My recommendation: if you smell electrical burning, turn off the lamp
and check your bulb!  If the smell persists, something else is amiss.  

And no one has mentioned benefits to running cooling fans across housing
this time around.  It seems to help extend lamp life.

Mike Ignatius

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Russell Spear
Sent: Friday, April 25, 2008 7:58 AM
To: [hidden email]
Subject: Re: HBO/Mercury lamp lifetimes

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I've been using HBO 100 W s, for about 30 years, and have had 2
explosions in that time, one long ago and one approx. 3 yrs past.  Both
could be attributed to a high number of on off cycles (something I
discourage).  I nomally cycle a bulb no more than twice a day, and need
to relamp about every 40-45 days in a busy lab with 3 fluorescent
scopes.  We always replace at 200 hrs. the cost of the bulb is small
compared to the damage to the equipment and time and cost of
decontaminating the microscope rooms.

Russ



Russell N. Spear
Sr. Research Specialist
Dept. of Plant Pathology
Univ. of Wisconsin
1630 Linden Dr.
Madison WI 53706

voice 608.263.2093
fax     608.263.2626
jlribas jlribas
Reply | Threaded
Open this post in threaded view
♦
♦
|

scanhead without stopping in time-lapse mode

In reply to this post by Russell Spear
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear Confocalists,

In time-lapse mode using a Leica TCS SP2 the scanhead does not stop
moving during the overall experiment. But the programed stacks are
adquired correctly so the laser(s) only reach the sample when it is the
time for the real adquisition. My question is if it is a normal feature
of the scanhead in Leica or if I have some problem with the unit. If it
is so, will a 48 hours time-lapse experiment be dangerous for the scanhead?
Thank you very much in advance

--
Juan Luis Ribas
Servicio de Microscopía
Centro de Investigación, Tecnología e Innovación
Universidad de Sevilla
Av. Reina Mercedes 4b
41012 Sevilla
Spain
Glen MacDonald-2 Glen MacDonald-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Ignatius, Mike
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We recently encountered a failure of a Nikon Xenon housing on a  
vintage 1992 Diaphot.  The hard plastic insulation on part of the  
cable interface broke down and arced to the lamp housing at  
ignition.  The users reported that it kept clicking (sparking),  
wouldn't ignite and the burning odor wouldn't go away.  Gee, at least  
they finally got the idea and turned it off.

Our only other non-bulb issue was a Prior power supply on a Leica  
MZ-12 FLIII that Prior never could get to keep igniting.

  I change our Hg bulbs about 300-320 hrs  on widefield systems, but  
let a confocal go to 350-380 since it has fewer ignitions.  The other  
confocal has a 75W xenon with about 4,000 hours.

Ushio 102D bulbs have been much more reliable than the Osram 103W,  
especially with older power supplies.    I've replaced all of the  
Osram bulbs with Ushio.

Glen

Glen MacDonald
Core for Communication Research
Virginia Merrill Bloedel Hearing Research Center
Box 357923
University of Washington
Seattle, WA 98195-7923  USA
(206) 616-4156
[hidden email]

************************************************************************
******
The box said "Requires WindowsXP or better", so I bought a Macintosh.
************************************************************************
******


On Apr 25, 2008, at 10:41 AM, Ignatius, Mike wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> An important side bar in this discussion, is what warning signs are
> there that the bulb is too old and possibly going to fail/explode.
>
> For me, the earliest signs that the arc gap has grown to impossible
> widths were difficulty in starting the bulb to begin with, flickering
> when on, long time to stabilize etc.
>
> The more dreadful indication is an electrical burning smell when it is
> on and about to blow.
>
> In my one experience it smelled like insulation burning off of wiring.
> Apparently the bulb super heats before failure?  I would be curious to
> hear if others have had this warning sign before the bulb blew.  I
> stupidly got up, and sniffed around the scope to see what the  
> source was
> and wham the bulb blew in my face.  I thanked Nikon for their sturdy
> housing, no glass came through, though the diffuser lens inside did
> shatter, and the mercury vapor surely escaped.  After it has blown,
> always good to leave the area for a while to let the gas escape.
>
> My recommendation: if you smell electrical burning, turn off the lamp
> and check your bulb!  If the smell persists, something else is amiss.
>
> And no one has mentioned benefits to running cooling fans across  
> housing
> this time around.  It seems to help extend lamp life.
>
> Mike Ignatius
>
> -----Original Message-----
> From: Confocal Microscopy List  
> [mailto:[hidden email]] On
> Behalf Of Russell Spear
> Sent: Friday, April 25, 2008 7:58 AM
> To: [hidden email]
> Subject: Re: HBO/Mercury lamp lifetimes
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> I've been using HBO 100 W s, for about 30 years, and have had 2
> explosions in that time, one long ago and one approx. 3 yrs past.  
> Both
> could be attributed to a high number of on off cycles (something I
> discourage).  I nomally cycle a bulb no more than twice a day, and  
> need
> to relamp about every 40-45 days in a busy lab with 3 fluorescent
> scopes.  We always replace at 200 hrs. the cost of the bulb is small
> compared to the damage to the equipment and time and cost of
> decontaminating the microscope rooms.
>
> Russ
>
>
>
> Russell N. Spear
> Sr. Research Specialist
> Dept. of Plant Pathology
> Univ. of Wisconsin
> 1630 Linden Dr.
> Madison WI 53706
>
> voice 608.263.2093
> fax     608.263.2626
Dale Callaham Dale Callaham
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: HBO/Mercury lamp lifetimes

In reply to this post by Ignatius, Mike
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,

I have seen strong recommendations against forced air cooling where arc
stability is a prime concern. Osram literature states that forced air is
not needed if sufficient convection through the lamp housing will keep
the bases below 230C and most commercial designs probably do this. They
state that any cooling should "not be directed at the discharge vessel".
Most recent lamp housings and lamp mountings assist the upper base
cooling with a heatsink. I bolted a thermocouple to the upper base (-
terminal) of a unit (leave TC disconnected until after ignition!) and
measured only 180C for a new lamp; the lamp has to get hot to function
properly. Staying with well-designed equipment in good operating
condition (check for oxidation of leads and connectors and replace if
necessary) it shouldn't be necessary to use forced air.

I am curious - someone mentioned the Zeiss AttoArc - doesn't that allow
you to run at lower than max/nominal settings? Seems that many arclamps
are too bright and need attenuation anyway; I was wondering if anyone
runs them at lower levels as a standby condition between users or for
lower excitation and does it seem to extend the life? I built a unit
that could run a HBO 100W/2 nicely below 1A (they have a 5A nominal
current) but I didn't get to monitor the performance over a long period.

Dale

Ignatius, Mike wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> An important side bar in this discussion, is what warning signs are
> there that the bulb is too old and possibly going to fail/explode.  
>
> For me, the earliest signs that the arc gap has grown to impossible
> widths were difficulty in starting the bulb to begin with, flickering
> when on, long time to stabilize etc.  
>
> The more dreadful indication is an electrical burning smell when it is
> on and about to blow.
>
> In my one experience it smelled like insulation burning off of wiring.
> Apparently the bulb super heats before failure?  I would be curious to
> hear if others have had this warning sign before the bulb blew.  I
> stupidly got up, and sniffed around the scope to see what the source was
> and wham the bulb blew in my face.  I thanked Nikon for their sturdy
> housing, no glass came through, though the diffuser lens inside did
> shatter, and the mercury vapor surely escaped.  After it has blown,
> always good to leave the area for a while to let the gas escape.  
>
> My recommendation: if you smell electrical burning, turn off the lamp
> and check your bulb!  If the smell persists, something else is amiss.  
>
> And no one has mentioned benefits to running cooling fans across housing
> this time around.  It seems to help extend lamp life.
>
> Mike Ignatius
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of Russell Spear
> Sent: Friday, April 25, 2008 7:58 AM
> To: [hidden email]
> Subject: Re: HBO/Mercury lamp lifetimes
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> I've been using HBO 100 W s, for about 30 years, and have had 2
> explosions in that time, one long ago and one approx. 3 yrs past.  Both
> could be attributed to a high number of on off cycles (something I
> discourage).  I nomally cycle a bulb no more than twice a day, and need
> to relamp about every 40-45 days in a busy lab with 3 fluorescent
> scopes.  We always replace at 200 hrs. the cost of the bulb is small
> compared to the damage to the equipment and time and cost of
> decontaminating the microscope rooms.
>
> Russ
>
>
>
> Russell N. Spear
> Sr. Research Specialist
> Dept. of Plant Pathology
> Univ. of Wisconsin
> 1630 Linden Dr.
> Madison WI 53706
>
> voice 608.263.2093
> fax     608.263.2626